Thread with our findings (conducted by @petrosoniak & @HumanFact0rz )
We used a modified pluralistic walkthrough technique
@NicholasChrimes
1/
We identified multiple logistical and ergonomic issues associated with the use of the intubation box & based on this WE CANNOT RECOMMEND this device at this time within an emergency department setting.
For more details, see below
2/
3/
1. pre-intubation
2. peri-intubation
3. post-intubation
4/
This device is large and requires a dedicated storage location
A cleaning process is needed to ensure consistently ready for use
5/
A variety of our staff tested and were able to transport it to the bedside without assistance
Our taller staff could place on beds but our shorter staff, found it challenging requiring more help
6/
We piloted w/ 3 intubators and optimal bed position
Short (5ft1in): visualization OK
Average (5ft7in): Problematic
Tall (6ft2in): Adequate
Glare from overhead lights is significant for tall operators & impacted visuals during initial steps of intubation
9/
10/
There were delays noted in the intubation process of at least 30 seconds.
12/