Ever wonder how dialysis units make that fancy, high-quality water? It has always interested me as a nephrologist. Here's a 🧵with photos👇 1/12
⭐️Potable water💧enters the treatment circuit and the mixing valve blends to the desired temperature
⭐️The valve is equipped with a thermometer🌡️and adjustment knob
2/12
⭐️A backflow assembly prevents the dialysis water from entering back into the city water supply
3/12
⭐️A booster pump helps to provide adequate/continuous water pressure for operation💪
⭐️Also includes an electric pump controller (equipped with a pressure transducer) and expansion tank
⭐️The expansion tank helps smooth out pressure changes
4/12
⭐️ ON THE LEFT - A depth filter removes coarse particulate matter. The filter consists of coal, sand, and garnet as the filtration media
⭐️ ON THE RIGHT - A water softener reduces levels of calcium and magnesium.
⭐️Ca and Mg ions are exchanged for Na
5/12
⭐️Softeners have a finite capacity and must be regenerated
⭐️Regeneration is completed by exposing the resin to a strong brine solution
6/12
⭐️Carbon tanks are next in line!
⭐️They are the principal method to remove chlorine and monochloramine, substances that can degrade reverse osmosis membranes and harm patients
7/12
⭐️ A backup carbon filter is often used in series
⭐️Carbon beds are closed monitored to detect exhaustion of their removal capacity
⭐️Frequent testing is necessary to verify that total chlorine levels remain ≤0.1 parts per million (PPM)
8/12
⭐️Just before water reaches the Reverse Osmosis unit, a pre-filter clears out any particulate matter that may damage the membrane
9/12
⭐️Reverse osmosis (RO) removes substances such as metal ions/salts, chemicals, and microbial elements (bacteria, endotoxins, and viruses)
⭐️Organic contaminant removal is largely based on size; those with a molecular weight >200 Da are filtered
⭐️Uses Hydrostatic pressure!
10/12
⭐️Deionization removes ionic contaminants when the RO fails or when additional purification is necessary
⭐️H+ exchanged for➡️Cations
⭐️OH- exchanged for➡️Anions
⭐️H+ and OH- combine to become water
⭐️Followed by an endotoxin filter since DI systems promote bacterial growth
11/12
So:
✅ Dialysis patients are exposed to ~150L of dialysis solution each treatment
✅ Providing clean water for dialysis is critical for patient safety and positive patient outcomes
✅ The components of H20 treatment need to be monitored closely to ensure optimal performance
Fin
• • •
Missing some Tweet in this thread? You can try to
force a refresh
A case of Metformin-induced lactic acidosis (MALA) including a brief thread covering risk factors, kinetics, and indications for hemodialysis 🧵 1/6
Patient with T2DM and chronic alcoholism presented with 2 weeks of poor PO intake, N/V, abdominal pain, and diarrhea. ⬇️BP +⬆️RR in the ER
Risk factors for MALA:
🔸CKD
🔸Liver Disease
🔸Active Alcohol Use
🔸Heart Failure
🔸Series Acute Illness
🔸Hemodynamic Instability
Mechanism of MALA
🔹Metformin promotes conversion from glucose ➡️ lactate in the splanchnic bed of the small intestine
🔹 Also inhibits mitochondrial respiratory chain complex 1 leading to ⬇️ hepatic gluconeogenesis from lactate and pyruvate. This = additional lactate/substrate