Nick Mark MD Profile picture
Aug 21, 2022 26 tweets 11 min read Read on X
"Hey what’s this funny donut 🍩 stuck on the crash cart?"

It's a 🧲 magnet! We keep it there because it can be life-saving if applied over a pacemaker or ICD.

A 🧵about magnets, pacemakers & ICDs. Some useful facts that everyone in an ICU ought to know.
#FOAMcc #FOAMed
1/
Image
Image
(re-tweeting the thread to fix a few minor errors)

Pre-tweetorial question:
What happens when you place a magnet on an ICD?

* note that the results vary by device programming & manufacturer (more on this later)

2/
The answer is A&B. Unlike applying a magnet 🧲 to a pacemaker, for most ICDs applying a magnet only stops the anti-tachycardia therapies.

More on the details later.

But first a quick review on pacer/ICDs.

3/
Pacemakers electrically *stimulate* either the atria, ventricle, or both.
Most also *sense* the patients rhythm. This way they avoid pacing on top of the person’s native beats.
4/
In contrast, Implantable Cardioverter Defibrillators (ICDs) *sense* the rhythm & respond to arrhythmias.
When ICDs sense a life-threatening arrhythmia (VT, VF) they can *cardiovert/defibrillate* or perform *anti-tachycardia pacing*.
All ICDs can also function as pacemakers.
5/
You can usually spot an ICD on CXR because the leads have thick “shock coils" for cardioversion.

(If you are wondering why ICDs have “shock coils" it’s to distribute energy over a larger area to avoid tissue damage. Just like external defib pads!)
pubs.rsna.org/doi/epdf/10.11…
6/
Pacemakers & ICDs are lifesaving, but occasionally they malfunction (broken lead, trouble sensing, etc).

There’s also certain circumstances (electrosurgery, EOL, etc) where we need to disable certain device functions.

That’s where 🧲 magnets & simple electronics come in!
7/
Before we get to what happens, how does the pacemaker/ICD know that a magnet has been applied?

One of two technologies:
-a Reed switch (older devices)
-a Hall effect sensor (newer devices)

8/
The first - a Reed switch - is a super simple device:
It's just two flexible wires that will close a circuit if a magnetic field is applied.
Doesn't get much simpler than that!
9/
If you look really closely at an older pacemaker you can actually see the Reed switch:
10/
Fun fact: You have almost certainly interacted with a Reed Switch today. You just didn’t know it.
Here are very two common household examples:
11/
More modern pacemakers/ICDs (including all MRI conditional ones) use a slightly more sophisticated device: a Hall Effect sensor.

Hall Effect sensors are more sensitive & output a voltage proportional to the magnetic field strength.
This is more info than just open/close.
12/
Like a Reed Switch, you’ve almost certainly interacted with a Hall effect sensor today.
They are used to measure the RPMs on rotating parts. For example in a car transmission or an exercise bike . (Yes there’s 1 in a peloton!)
They also provide the compass in smart phones.
13/
So now that we know how implanted devices sense magnets. What happens if we put a magnet on a pacemaker?

Here’s a great example from ECGpedia.
en.ecgpedia.org/wiki/McGill_Ca…
13/
Applying a magnet to most* pacemakers puts them an asynchronous mode (e.g. it stops sensing) but continues pacing at a fixed rate.

Specifically, it will put a pacer into AOO/VOO/DOO mode (depending on if what leads are present).

(*note: some not programmed this way!)
14/
Why would we want to do this?
Inappropriate pacing can cause Pacemaker mediated tachycardia (AKA "endless loop tachycardia") wherein the pacer V lead fires, senses a retrograde atrial conduction, & fires again... endlessly.
Applying a magnet interrupts this endless loop!
15/
Another situation occasionally encountered in the OR, is that a pacemaker may respond inappropriate to electrocautery pulses.

Usually these pulse are mistook for native complexes and the device UNDER-paces as a result, potentially causing bradycardia
ahajournals.org/doi/10.1161/CI…
16/
Interestingly, with a 🧲 magnet applied, some older pacemakers will tell you it's battery status:
For example a BS pacer will beat at 100 bpm if the battery is full🔋& 85 bpm if low 🪫
(beginning of life=BOL, elective replacement=ERT)

See this algo👇
academic.oup.com/europace/artic…
17/
What about putting a magnet on an ICD?

The details vary by manufacturer, but in general it will:
- disable all tachy-therapies (pacing, cardioversion, & defibrillation)
- on most devices it does NOT effect pacemaker functions (unlike a pacemaker)
19/
Note that this is an important contrast between ICDs and pacemakers - and something I did a really bad job explaining in an earlier version of this thread!

Many thanks to the courageous EP and cardiologist of twitter who educated me
20/
But why would you ever want to turn off a ICD’s anti-tachycardia function?

Three reasons:
-to stop inappropriate shocks from being delivered
-to prevent the ICD from firing inappropriately due to electrocautery
-to avoid painful shocks as part of DNR/comfort care
21/
For example a fractured ICD lead can cause improper sensing leading to multiple inappropriate shocks ⚡. Ouch!

Applying a ring 🧲 magnet will temporarily stop the ICD from delivering more unnecessary shocks. (this person will need the lead replaced)
cardiocases.com/en/pacingdefib…
22/
Another situation we sometimes see in the ICU is ICDs at the end of life.
Defibrillation by ICD is generally inconsistent with the goals of people who are DNR.
Applying a magnet can help avoid the anguish of futile shocks in a dying person.

See here: mypcnow.org/fast-fact/impl…
23/
Note that using a magnet for ICD deactivation at EOL is really only a stopgap measure.
A better solution to taping a magnet to someone's chest is to have the device rep come in and deactivate the ICD. Depending on where you work this can be pretty quick (hours).
24/
To summarize, we learned:
- how pacemakers & ICDs respond to magnets (& how it's different!)
- what happens when a magnet is applied
- specific times that is useful to apply a magnet
(bonus: you even learned about magnetic sensors in everyday devices!)

Hope you enjoyed!
25/
Ok a few more bonus 🧲 facts:
Not quite *every* pacemaker is magnet responsive.
The newest & smallest pacemakers, which are implanted via catheter directly in the ventricular wall, do NOT respond to an external magnet.
ncbi.nlm.nih.gov/pmc/articles/P…
24/

• • •

Missing some Tweet in this thread? You can try to force a refresh
 

Keep Current with Nick Mark MD

Nick Mark MD Profile picture

Stay in touch and get notified when new unrolls are available from this author!

Read all threads

This Thread may be Removed Anytime!

PDF

Twitter may remove this content at anytime! Save it as PDF for later use!

Try unrolling a thread yourself!

how to unroll video
  1. Follow @ThreadReaderApp to mention us!

  2. From a Twitter thread mention us with a keyword "unroll"
@threadreaderapp unroll

Practice here first or read more on our help page!

More from @nickmmark

Oct 13
#HurricaneHelene damaged the factory responsible for manufacturing over 60% of all IV fluids used in the US, leading to a major national shortage.

As clinicians what can we do to about the #IVFluidShortage and how can we prevent this crisis from happening again?

A thread 🧵
1/ Image
There are many things we can do as clinicians to improve ICU care & reduce IVF use.

1️⃣Don't order Maintenance IV Fluid!
Almost no patient actually needs continuous IV fluids.
Most either need resuscitation (e.g. boluses) or can take fluid other ways (PO, feeding tube, TPN).
2/
Frequently if someone is NPO overnight for a procedure, MIVF are ordered.
This is wrong for two reasons.
We are all NPO while asleep & don't need salt water infusions!
We should be letting people drink clears up to TWO HOURS before surgery, per ASA.

3/ pubs.asahq.org/anesthesiology…Image
Read 16 tweets
Oct 1
New favorite physiology paper: Central Venous Pressure in Space.

So much space & cardio physiology to unpack here including:
- effects of posture, 3g shuttle launch, & microgravity on CVP
- change in the relationship between filling pressure (CVP) & LV size
- Guyton curves!
1/

Image
Image
Image
To measure CVP in space they needed two things:
📼 an instrument/recorder that could accurately measure pressure despite g-force, vibration, & changes in pressure. They built & tested one!
🧑‍🚀👩‍🚀👨‍🚀 an astronaut willing to fly into space with a central line! 3 volunteered!
2/
Image
Image
The night before launch they placed a 4Fr central line in the median cubital vein & advanced under fluoro.

🚀The astronauts wore the data recorder under their flight suit during launch.

🌍The collected data from launch up to 48 hrs in orbit.
3/
Image
Image
Read 16 tweets
Jul 16
The media silence on this is deafening.

Did he have a head CT? What did it show?
Did he have stitches? Tetanus shot?

The NYT ran nonstop stories about Biden’s health after the debate but can’t be bothered to report on the health of someone who was literally shot in the head?
To the people in the replies who say it’s impossible because of “HIPPA”
1. I assume you mean HIPAA
2. A normal presidential candidate would allow his doctors to release the info. This is exactly what happened when Reagan survived an assassination attempt.
washingtonpost.com/obituaries/202…
My advice to journalists is to lookup tangential gunshot wounds (TGSW).

Ask questions like:
- what imaging has he had?
- what cognitive assessments?
- has he seen a neurosurgeon or neurologist?
- he’s previously had symptoms like slurred speech, abnormal gait - are these worse?
Read 4 tweets
Jun 30
You've probably heard "don't give lactated ringers because it raises lactate"

This statement is ~98% false, but there's one crucial practice-changing fact that you need to know.

A 🧵 all about lactic acid and lactated ringers!
1/
Image
Image
First off, we should ackowledge the obvious: Lactated ringers does in fact contain lactate... 28 mEq/L in fact

BUT there's one little detail to remember:
Lactate ≠ Lactic acid

When we measure "lactate" we care about the ACID (H+) which lowers pH & causes organ dysfunction
2/ Image
But the correlation between pH & lactate is really bad!

Look at this analysis of lactate vs pH in 171 ICU patients.

There is a *weak* correlation in people with arterial lactate > 5, but even w/ lactate =10, pH ranged from 7.5 to 7.05. Quite a spread!

3/ ncbi.nlm.nih.gov/pmc/articles/P…
Image
Read 10 tweets
Jun 13
If you intubate you need to read the #PREOXI trial!
-n=1301 people requiring intubation in ED/ ICU were randomized to preoxygenation with oxygen mask vs non-invasive ventilation (NIV)
-NIV HALVED the risk of hypoxemia: 9 vs 18%
-NIV reduced mortality: 0.2% vs 1.1%

#CCR24
🧵
1/

Image
Image
Image
Hypoxemia (SpO2 <85%) occurs in 10-20% of ED & ICU intubations.

1-2% of intubations performed in ED/ICU result in cardiac arrest!

This is an exceptionally dangerous procedure and preoxygenation is essential to keep patients safe.

But what’s the *BEST* way to preoxygenate?
2/
Most people use a non-rebreather oxygen mask, but because of its loose fit it often delivers much less than 100% FiO2.

NIV (“BiPAP”) delivers a higher FiO2 because of its tight fit. It also delivers PEEP & achieves a higher mean airway pressure which is theoretically helpful!
3/
Image
Image
Read 15 tweets
Jun 12
Results from #PROTECTION presented #CCR24 & published @NEJM.
- DB RCT of amino acid infusion vs placebo in n=3511 people undergoing cardiac surgery w/ bypass.
- Reduced incidence of AKI (26.9% vs 31.7% NNT=20) & need for RRT (1.4% vs 1.9% NNT=200)

Potential game changer!

🧵
1/
Image
Image
I work in a busy CVICU & I often see AKI following cardiac surgery.

Despite risk stratification & hemodynamic optimization, AKI remains one of the most common complications after cardiac surgery with bypass.

Even a modest reduction in AKI/CRRT would be great for my patients.
2/ From Nature reviews nephrology  https://www.nature.com/articles/nrneph.2017.119
During cardiac surgery w/ bypass, renal blood flow (RBF) is reduced dramatically. This causes injury, especially in susceptible individuals.

But what if we could use physiology to protect the kidneys?

Renal blood vessels dilate after a high protein meal increasing RBF & GFR!
3/ https://www.jtcvs.org/article/S0022-5223(18)33243-4/fulltext
Read 11 tweets

Did Thread Reader help you today?

Support us! We are indie developers!


This site is made by just two indie developers on a laptop doing marketing, support and development! Read more about the story.

Become a Premium Member ($3/month or $30/year) and get exclusive features!

Become Premium

Don't want to be a Premium member but still want to support us?

Make a small donation by buying us coffee ($5) or help with server cost ($10)

Donate via Paypal

Or Donate anonymously using crypto!

Ethereum

0xfe58350B80634f60Fa6Dc149a72b4DFbc17D341E copy

Bitcoin

3ATGMxNzCUFzxpMCHL5sWSt4DVtS8UqXpi copy

Thank you for your support!

Follow Us!

:(