Here I will explain what is meant by 'French' when we are talking about Foleys and NG tubes and such.
Also we will cover what the 'Gauge' of an IV means.
(1/ )
'French' isn't named after France 🇫🇷, the country.
It's (indirectly) named after a person -- Joseph-Frédéric-Benoît Charrière, a Swiss instrument maker who started off making knives but went on to make a huge variety of surgical instruments.
When Charrière made instruments, he picked his own unit of measurement that corresponded to 1/3 of a millimeter...this was the 'Charrière'.
But English speakers struggled to pronounce 'Charrière', so the companies changed it to 'French'.
So 1 'French' became 1/3 millimeter.
Again: 1 Fr = 1/3 mm.
Only things that are circular tubes have a 'French'. So things like central lines, NG tubes, Foleys, etc.
The 'French' always refers to the *outer diameter* of a tube.
For example, an 18 Fr NG tube = an NG tube with an *outer diameter* of 6 mm.
Here is a 7 Fr central line (triple lumen).
Again, the '7 Fr' refers to the *outer diameter*. It doesn't matter how many lumens it may have: that's a separate issue. It's 7 Fr, and therefore its outer diameter is 7/3 = 2.33 mm.
It's the same unit of measurement for all devices:
A 12 Fr NG tube has the same outer diameter as a 12 Fr dialysis catheter, a 12 Fr Foley catheter, or a 12 Fr chest tube.
All of these 12 Fr devices will have an outer diameter of 12/3 = 4 mm.
In some countries that have less of a problem pronouncing 'Charrière', catheters instead use the 'Ch' abbreviation instead of 'French'.
Here we see an '18 Ch' Foley, and we are reminded again that this means it's 6 mm in diameter.
On the other hand 'Gauge' is a different thing entirely.
'Gauge' refers to the Birmingham Gauge system, devised in the 1900's in England for manufacturing metal wire. It later became used for IV catheters.
There are a couple of major differences...
First, as the 'Gauge' number gets *larger*, the IV's get *smaller*. This is the opposite of the 'French' system.
As you can see here, a 14 Fr IV catheter is much larger than a 24 Fr catheter.
Also the 'Gauge' sizes are somewhat arbitrary...meaning the spacing isn't mathematically consistent.
So a 10 Gauge IV is *not* twice the size of a 20 Gauge IV, for example.
If you do the math here, you see these relationships are not linear or consistent.
Finally, as a trauma surgeon, I have to point out one of my favorite teaching points:
The flow rate of a 16 G IV is **twice** that of an 18 G IV. So the difference between those two IV's may not sound much at first, but it makes a big difference for major resuscitations...
⬛️
Of course I have a critical typo in my most-read 🧵😬😤😫
As the picture indicates, these IV's are expressed in terms of "Gauge". A couple of slides ago, I accidentally used the wrong abbreviation in the text.
Also, not to complicate things, but some have pointed out that there’s an exception for vascular access sheaths.
Here the companies have chosen to use “French” to indicate the size of the things that can pass through them.
🧵regarding some technical aspects of the needle holder and where to grip surgical needles, something you may not know about the packages they come in, and a comment about the ratchet.
Oriented more toward the younger folks... (1/ )
A good general rule is that the needle should be gripped about 2 mm from the tip of the needle holder.
This location usually provides enough strength to the grip and allows for the most freedom of motion of the needle holder.
Of course there's nothing preventing you from gripping it more proximal -- say 4 mm. And indeed that does provide a stronger grip (it's been studied).
The downsides are:
- the needle is more likely to bend
and
- the jaw tips protrude more and may hit things and get in the way.
Redoing one of my very first 🧵for followers that didn't see it the first time...
Short🧵 about what it means when an instrument has a gold handle.
The gold handles look cool but are the instruments actually any better? Read on to find out...
(1/ )
For needle holders, the gold handle means that the holder has tungsten carbide inserts.
Tungsten carbide is harder than stainless steel, so the jaws can have more teeth and the teeth are more prominent.
This enables it to hold onto the needle better and keeps it from moving.
Here you can see a closeup of the inserts. Again, the increased hardness of tungsten carbide allows for sharper teeth and more teeth, and the teeth are more resistant to wear.
Surgical needles are much less likely to slip or rotate while sewing, though they may bend more easily.
🧵regarding one of the most basic and commonly used instruments in surgery...the DeBakey forceps.
DeBakey forceps were designed for vascular and cardiac surgery, but now are used by almost everyone.
I'm surprised myself that we haven't covered it already.
So here goes... 1/
Michael DeBakey doesn't need much introduction: he was one of the most famous surgeons of all time. He was a pioneer in cardiothoracic and vascular surgery and was at Baylor for decades, eventually becoming chancellor. His rivalry with Denton Cooley was the stuff of legends.
DeBakey also was by all accounts very malignant, even when judged against old-school cardiac surgeons in Texas. And other behaviors were problematic. He wouldn't last 24 hr at a US hospital today.
He leaves a large but complex legacy that everyone has their own opinion about.
Today's short 🧵 is on the Hoffman clamp, a device that may only be of interest to a few of you...
But I picked it because I enjoy pointing out when simple devices play a key role in complex situations.
Here, we see a Hoffman clamp restricting flow in an ECMO circuit. (1/ )
... but why would you ever want to use a clamp to partially occlude the tubing, you ask??
After all, the ECMO machine has a knob that controls flow, does it not? Can't you just turn the knob to dial the flow down? 🤔
But it's different for more complex circuits.
This patient is on VAV ECMO. The drainage cannula is in the femoral vein...but now there are 2 'return' cannulas...one in the femoral artery and also one in the RIJ vein. They are joined by a Y connector. The circuit is 'VAV'.