Rawan Amir Profile picture
Feb 1 27 tweets 7 min read
1/🏥It's your first night alone in the CCU
😵‍💫You admit a patient with syncope due to complete heart block
🕡HR is 30 bpm with a wide ventricular escape rhythm
You wonder if it's time for your first #TVP, but
🔆is it indicated?
🔆& how are you going to set it up for success?
2/ Learning Objectives:
🔆Recognize indications for pacing
🔆How to manage TVP/adjust settings
3/ But first let’s start with a huge thank you to @AmitGoyalMG @A_h_ghoneem and @RichardAFerraro for their never-ending guidance and help making this happen! 👏🏼

@CardioNerds 🫀
4/ The author has no conflicts of interest.  
There is no commercial support for this content.
5/ Let's first take a vote 🗳️

I feel comfortable identifying patients who may require a TVP and troubleshooting common TVP problems
6/ What are the most common indications for TVP? 🤔

🔆Bradyarrhythmia à symptoms😵‍💫or hemodynamic compromise 🏥(eg. sinus bradycardia, 2nd or 3rd degree AV block)
🔆Overdrive pacing (eg. VT, TdP)
🔆Digoxin toxicity💊
🔆PPM Failure

ncbi.nlm.nih.gov/pmc/articles/P…
7/ So let’s get to know our TVP settings better🤓

For our case, our TVP mode will be VVI
V=Ventricle sensed
V=Ventricle paced
I=pacing Inhibited if native beat is sensed

Vs the other common mode of VOO
V=Ventricle paced
O=not sensed
O=no inhibition of pacing
8/ Although there are different types of generators, they all have the same 3 variables that you can control:

🕑Rate
⚡️Output
⚠️Sensitivity

Let’s start with Rate🕑meaning the number of impulses generated by your TVP/minute
9/ Next, let’s look at the Output ⚡️which is the current🔌produced by the generator during every beat (measured in milliAmpers, mA)

The⬆️the mA, the⬆️the output

(obvious right? wait until you see what's coming next though😉)
10/ Finally, let’s talk about Sensitivity ⚠️which is the minimum current 🔌needed to detect native cardiac activity and inhibit pacing 🫀(measured in millivolts, mV)

🚨Don’t be fooled!

The⬇️the mV, the ⬇️ current is needed to be detected by the TVP so the⬆️the sensitivity!
11/ To make sure we really understand the concept of sensitivity ⚠️

Let’s look at our favorite sensitivity analogy where you’re the pacer trying to see the QRS complex from behind a fence 👇🏻

Fence = mV
12/ 🚨So let’s go back to the CCU!

Your patient now has a TVP, rate is set at 60 bpm, output 10, sensitivity 8, but you see this on tele 👇🏻

So,
What is the problem?🤔
What are potential causes? 📋
How can you troubleshoot it? 🔧
13/ 🚨Prob = undersensing ⚠️

Pacer doesn’t detect intracardiac signal🫀(native QRS) ➡️asynchronous pacing

Possible causes➡️low sensitivity (fence too high), electrolyte abnormality, lead failure
14/ ✅ How to correct?

⬆️sensitivity by ⬇️mV, this allows pacer to detect more cardiac signals🫀

Review reversible causes such as electrolytes🩸

Check hardware (lead placement)🛠️
15/ You ⬆️sensitivity from 8 mV to 7 mV ➡️you are no longer seeing asynchronous pacing (no inappropriate pacing spikes after native QRS)

You cut down mV to roughly half (7 mV ➡️4 mV) and perfecto!👌🏼

Your patient is being paced at a rate of 60 bpm and feels great!👏🏼
16/ Congratulations! First curveball dodged!

But now, your patient’s tele shows this 👇🏻

So, let’s try this again

What is the problem?🤔
What are potential causes? 📋
How can you troubleshoot it? 🔧
17/ 🚨Prob = capture 🥅 failure

An impulse is generated by pacer but it does not cause myocardial depolarization🫀
(pacing spikes not followed by QRS complexes)

Possible causes➡️lead dislodgment, low output
18/ ✅ How to correct?

Increase output⚡️by increasing mA

Check lead positioning
19/You ⬆️ output from 10 mA to 12 mA ➡️ each pacing spike is now followed by a QRS complex (capture achieved! 👏🏼)

You double your output to 24 mA, and the patient feels great! 🥳

But why do you need such a high output? 🤔

Check your lead position, it probably moved!
20/ Amazing job so far! But wait, you now look at tele and you see this 👇🏻
Your patient is in complete heart block again! And there are no pacing spikes at all! 💔
So, one last time folks 🦾
What is the problem?🤔
What are potential causes? 📋
How can you troubleshoot it? 🔧
21/ it could be one of two problems:

DDx 1: Oversensing ⚠️

Pacer senses signals it shouldn’t (eg. T waves or P waves) thinking they are QRS complexes ➡️ inappropriate inhibition of pacing

Possible causes➡️ high sensitivity (fence too low), lead failure
22/ 🚨DDx 2: output⚡️failure

Native HR < set rate🕑, but still no pacing
(no pacing spikes when there should be pacing spikes)

Possible causes ➡️lead failure, generator failure, battery failure, oversensing
23/ So how can you trouble shoot? 🔧

⬇️ sensitivity by ⬆️ mV (bring the fence up), will overcome both DDx!

Check hardware (as below) 🛠️
24/ General check list 📜to go through when troubleshooting🧑🏻‍🔧

-generator on? 🔛
-Battery low? 🪫
-Wires attached well?
-Leads connected?🔌
-Could the TVP wire have been displaced?
25/ So lets ask again now 👀

I feel comfortable identifying patients who may require a TVP and troubleshooting common TVP problems🦸🏾‍♀️
26/ I learned something in this Tweetorial that may change my clinical practice🤓
27/ Congratulations on making it to the end of this thread and for keeping your patient alive in the CCU! 🥳

Thank you so much for listening and hope you learned something new from this little tweetorial 📚

#Cardiotwitter #TVP

• • •

Missing some Tweet in this thread? You can try to force a refresh
 

Keep Current with Rawan Amir

Rawan Amir Profile picture

Stay in touch and get notified when new unrolls are available from this author!

Read all threads

This Thread may be Removed Anytime!

PDF

Twitter may remove this content at anytime! Save it as PDF for later use!

Try unrolling a thread yourself!

how to unroll video
  1. Follow @ThreadReaderApp to mention us!

  2. From a Twitter thread mention us with a keyword "unroll"
@threadreaderapp unroll

Practice here first or read more on our help page!

Did Thread Reader help you today?

Support us! We are indie developers!


This site is made by just two indie developers on a laptop doing marketing, support and development! Read more about the story.

Become a Premium Member ($3/month or $30/year) and get exclusive features!

Become Premium

Don't want to be a Premium member but still want to support us?

Make a small donation by buying us coffee ($5) or help with server cost ($10)

Donate via Paypal

Or Donate anonymously using crypto!

Ethereum

0xfe58350B80634f60Fa6Dc149a72b4DFbc17D341E copy

Bitcoin

3ATGMxNzCUFzxpMCHL5sWSt4DVtS8UqXpi copy

Thank you for your support!

Follow Us on Twitter!

:(