1/ 🎯Bland-Altman analysis is the common way to evaluate AI in #EchoFirst. However, BA analyses can be tricky to interpret. 🧵On the paradox of BA analysis for AI in #EchoFirst with examples from our studies on #AutoMAPSE 📜
#Foamed #Cardiotwitter rdcu.be/dH8aR
2/ BA analysis is an important approach used for comparing 2⃣measuement methods for measuring the same unit, like #MAPSE (mm) and #AutoMAPSE (mm). ‼️Correlation (r) is inappropriate yet commonly used because it is seductive. Read the reasoning in their📜: doi.org/10.1016/S0140-…
3/ The output in BA analysis is 1⃣bias &2⃣limits of agreement (LOA). Bias is the avg difference across all paired measurements of manual MAPSE and autoMAPSE. LOA=+/- 2SD of the bias in the study. The bias and LOA is presented in a BA plot from 📜: rdcu.be/dH0mY
4/ Ideally, the bias is zero. Two scenarios can give zero bias. 1⃣: The 2 methods measure EXACTLY the same in all cases - this is never the case. 2⃣: The 2 methods measure differently, but the differences are random and therefore cancel out in the study - this is common and OK.
5/ If the bias is very different from zero (defined clinically), inspect the BA plot. If the data is evenly spread around the bias, then the bias is systematic and can be calibrated by subtraction the bias. If not, then the two methods may be different: doi.org/10.1034/j.1399…
6/ The LOA is "the 95% CI of the bias". Because two methods NEVER measure EXACTLY the same, LOA is never +\- zero. How wide LOA is acceptable requires judgement, and there are two common approaches for this: the usual way and the AI-in-#EchoFirst way.
7/ The usual way is by asking: "Does differences in measurement by one-sided LOA cause clinical issues?" If no, then the two methods are interchangeable. This approach works when the one of the methods is the gold standard as the differences are from errors in the new method.
8/ Unfortunately, there is no gold standard in #EchoFirst studies. In our studies on #AutoMAPSE, humans are both the reference method AND the method to beimproved. This is common but makes interpretation of LOA difficult; the LOA can be due to human error or errors in #AutoMAPSE.
9/ Error in either method of BA analysis will widen the LOA. If errors in the reference is unrecognised, the AI-method will be discarded despite being a promising method. This Editorial in @BJAJournals show how error in the refrence method affects LOA: doi.org/10.1093/bja/ae…
10/ How to overcome this issue of human error? One popular solution is to assess how much human error to expect by assessing interobs variability. If LOA between #AutoMAPSE and manual #MAPSE is similar for two humans, then #AutoMAPSE is probably OK.
11/ Another solution is a test-retest study design to assess precision. If precision of #AutoMAPSE is good and bias is zero, then a wide LOA is likely human error. We tried test-retest for our study, but the study design was suboptimal. Better results are coming soon.
12/ A third solution is to assess the two methods for predicting an outcome. If they are equal, but the AI is faster n' easier, then use the AI irrespective of LOA. This is also how a new method can eventually replace the current standard: doi.org/10.7326/0003-4…
13/ So, back to #AutoMAPSE vs. manual #MAPSE. The bias was near-zero and LOA from this study was -3.7 to 4.5 mm; too wide if the reference was a gold standard. However, it was human measurements. We must determine how much human error to expect. rdcu.be/dH8aR
14/ Interobs variability was assessed a priori and the LOA was similar to our results (-4.7 to 3.0 mm): 📜. Thus, the LOA between #AutoMAPSE and manual #MAPSE seems OK, and we concluded that #AutoMAPSE provides valid measurements overall. doi.org/10.1093/ehjimp…
15/ However, we reassessed the interobs between another pair of observers, and the LOA was narrower...this emphasises the variability of human measurements and the challenges of using BA analysis alone. More studies on #AutoMAPSE with data other than BA analysis is coming soon!
unroll @threadreaderapp
• • •
Missing some Tweet in this thread? You can try to
force a refresh
🧵1/
A very interesting RCT by @MirjanaCihoric & @BangFoss in @BJSurgery show the importance of dexamethasone (DXM) in acute abdominal surgery! DXM ⬇️inflammation,⬇️vasodilation,⬇️pressor dependency, and improves fluid balance and even 90-day ☠️ doi.org/10.1093/bjs/zn…
2/ DXM shows positive effects on various secondary endpoints with different levels of significance. Also, less fatigue, pain, immobilization and better peak exp flow.
3/ Interestingly, there were no more infections in the DXM-group; in fact, there were fewer infections of various subtypes, though statistically insignificant. However, DXM had more hyperglycemia.
1/ 🧵Can the pulmonary artery catheter (PAC) guide massive transfusion?
Key ideas:
1⃣Continuous SvO2 may be useful for optimizing O2-delivery.
2⃣Continuous SvO2 & CO aids in detecting distributive shock physiology.
#FOAMed #MedTwitter #CritCare #ICU
2/ 1⃣Optimizing DO2
The goal of transfusion is to improve O2-delivery. SvO2 is an excellent measure O2-delivery. PAC allows continuous SvO2 monitoring.
3/ 1⃣Optimizing DO2
🩸SvO2 > 65% during bleeding suggests a need for transfusion.
🩸SvO2 > 65% suggests that resuscitation is adequate, avoiding unnessecary transfusions and potentially preventing transfusion-associated complications.
2/
1⃣st report of TEE was in 1980, using M-mode from TGSAX during cardiac surgery. The rationale for this was: "Currently used monitoring techniques such as measurements of pressure and cardiac output provide an incomplete picture of LV function..." doi.org/10.1016/0002-9…
3/ A serious limitation is that all measurements required manual image acquisition and measurements.
Imagine using Korotkoff sounds for "monitoring" BP; manual methods requires too much time and attention to effectively serve as monitoring and guide patient management⏱️🧠
🧵1/ #AutoMAPSE had high feasibility for estimating MAPSE from ≥1⃣wall. Many worry that #MAPSE is limited by RWMA, such as acute MI - however, evidence suggests that regional MAPSE reflects global LV function.
2/ Several studies have shown that in AMI, MAPSE is ⬇️even in walls remote to MI. This was confirmed experimentally by Berg et al, in an experiment occluding the LAD in pigs and measuring #MAPSE by #WhyCMR.
First paper:
Berg et al: doi.org/10.1080/140174… doi.org/10.1152/japplp…
3/ ⬇️MAPSE in remote walls is explained through the distribution of afterload. For a specific region to shorten, that region must overcome the opposing afterload imposed by its neighboring region, which is shortening in the opposite direction.
🧵We have now refined AutoMAPSE.!
🔓
⬆️New paper on #autoMAPSE showing that the newest version achieves an optimal balance between feasibility, analysis speed, and agreement.
A🧵on the clinical importance of these findings
#echofirst #CardioTwitter #POCUS doi.org/10.1016/j.ultr…
1/ Our design involved measurements from 🔟beats per #EchoFirst recording. This is very different from the usual practice of measuring 1-3 beats. What are the advantages?
2/ Measuring 1-3 beats in #Echofirst involves huge physiological variability. How?
❤️🫁 interactions can cause serious beat-to-beat variations in any echo measurement. Often overlooked, these changes may be misinterpreted as real. Acting on wrong info ⬆️the risk of harm.
The RVOT or PA Doppler has a lot of underutilized potential for hemodynamic monitoring of the RV #ThePeoplesVentricle
🧵On the physiology of the RVOT/PA Doppler. Simplified understandings. Corrections are welcome!😃
#FOAMcc #EchoFirst #Cardiotwitter #FOAMed #Medtwitter #POCUS
VTI reflects stroke volume, but values are much lower than for LVOT. Lower cut-off for 60-79 year olds is actually around 10 cm.
Monitoring changes are likely more useful. doi.org/10.1016/j.jcmg…
AcT and Vmax/AT both reflects RV afterload. Increased afterload reduces AT and Vmax/AT.
➡️AcT correlates with mPAP and PVR. doi.org/10.1161/01.CIR…