#IssaTweetorials
#EPeeps
Impedance and RF ablation:
Part 2: How does RF ablation affect impedance?
 
1/9
 As tissue temperature rises during RF energy application, ions within the tissue being heated become more mobile, resulting in a decrease in impedance to current flow. 
2/9
 There are currently 2 methods to measure impedance: Generator Impedance (GI) & Local Impedance (LI). 
3/9
 Lack of impedance drop during RF energy application can reflect inefficient energy delivery to the tissue due to poor tissue contact, lack of catheter stability, or inadequate power delivery. 
4/9
 LI is measured at the distal electrode of the catheter and represents near-field impedance at the ablation electrode-tissue interface. 
5/9
 LI drop rate and the magnitude of drop closely correlate to the intra-mural tissue temperature (2-4 mm below the tissue surface) during standard and high-power RF ablation and can thereby act as an intramural thermometer. 
6/9
 LI drop of 10-20-Ω is observed in clinically effective & safe (intramural temperature 55-90°C) RF applications. 
Larger LI drops (>35 Ω) often reflect excessive tissue heating (intramural peak temperature nearing 100°C), indicating a risk for steam pops/char formation. 
7/9
 LI offers 4-time greater working range than GI, which may permit more precise titration of energy delivery, especially during high-power RF ablation when tissue heating occurs rapidly. 
Compared to GI, LI offers more sensitive & more specific measure of intramural tissue heating, as well as larger and more rapid impedance drops during RF application, allowing LI to be used as a real-time monitor of RF lesion formation & to guide titration of RF energy delivery.
Share this Scrolly Tale with your friends.
A Scrolly Tale is a new way to read Twitter threads with a more visually immersive experience.  
        
Discover more beautiful Scrolly Tales like this.
