Important to note that the MHRA see these tests as for detecting “Current Infection”. That is detecting active infection, but is bigger than checking whether people are infectious. It also includes people who are pre-infectious.
2/12
They give guidance on reference standards to detect this. There are challenges here in establishing scientifically valid clinical reference standards. RT-PCR is regarded as acceptable which is what most are using.
3/12
Document gives “desirable” and “acceptable” performance levels, and separates out roles of the tests for RED LIGHT and GREEN LIGHT purposes, as they require different performance levels.
This all makes good sense.
So how well do they say these tests need to work?
4/12
Sensitivity needs to be >=80% for red light tests and 97% for green light tests, or even better.
5/12
Minimum sample size 150 cases, and evaluating the test in the intended use population.
Emphasises importance of empirical evaluation of strategies of tests, not just evaluating them in models.
Echoes many comments made by @RoyalStatSoc in report.
Notably these values are not reached by Innova – data are only available on 80 asymptomatic cases from Liverpool and Birmingham, with a much lower sensitivity and no data on self-testing.
7/12
Specificity requirements same for both uses, 99.5% acceptable 99.9% desirable. No problem for Innova on this one.
8/12
The guidance is very clear on the importance of strong study designs following the recommendations here. bmj.com/content/372/bm…
Great to see the regulator championing the importance of real world studies over analytical laboratory evaluations.
9/12
Interestingly that this means MHRA do not regard the laboratory based studies planned by the Test Valdiation Group as acceptable evidence.
There are some troubling disconnects between TVG and MHRA in these docs. TVG seems out of step.
They do acknowledge that more rapid study designs can be used in a pandemic, but insist on immediate commencement of methodological strong evaluations most-marketing
This is all very welcome
11/12
This seems well thought through and justified (there's lots of maths at the back) to explicitly stating the evidence and performance levels we need.
It may be some while before tests are available at these standards,
Daily Testing in school study report is out but presentation by BBC here is SPIN SPIN SPIN
The trial failed to show convincing reductions in school absence, and could not rule out large increases in Covid transmission. Sensitivity of the test was 53%.
BBC says that reduced absence by 33%, but the ITT analysis in the text says 20% reduction with 95% confidence interval from 46% reduction to a 19% increases (p=0.27). So no convincing evidence of a reduction.
Participants were first tested with LFTs – 810 positive and 1736 negative. The investigators choose to test 217 of the 1736 negatives with PCR – that’s 1 in 8. This wasn’t a random sample as they were influenced by clinical characteristics as well as the test result.
2/9
The sensitivity / specificity calculation is based on all LFT+ves and 12.5% of LFT-ves as follows:
Results from the LIVERPOOL EVENT PILOTS have been published on line and in the media. Somehow I missed these coming out. cultureliverpool.co.uk/event-research…
No official report from @dhscgov as per normal.
Seems important evidence is being delayed once again.
1/7
The bottom line is that the events were safe.
Kudos to Liverpool PH Team.
But detail is interesting to see why they were safe.
2/7
First the infection rate in Liverpool was very low when the events were held
Negative LFTs required for entry. 5/13263 positive and excluded. Same-day PCR found 4 people positive who had attended with false negative LFTs. So 5/9 were picked up by LFT – 44% missed.
3/7
What do we known about ORIENT GENE used in the Daily Contact Testing Trial by the @educationgovuk and @DHSCgovuk?
There have been claims that this test is as good as others and has been reviewed by @MHRAgovuk for use in assisted testing. This is not right
1/10
The process does not make sense.
The MHRA never review products for assisted testing as they are professional use tests, which go through the self-certification process to get a CE-IVD mark.
MHRA doesn't go near this process.
2/n
In fact ORIENT GENE is not even on the MHRA register of products which is a requirement. You can check here - both for the product and manufacturer (sorry for the messy link).