Q: If you could deliver 30 W of RF energy for 30 sec using any of the ablation (abl) electrodes shown in the figure, which RF ablation catheter creates larger ablation lesion size?
A: Let’s talk about how the RF abl lesion is formed.
2/10
The size of the lesion created by RF is determined by the amount of tissue heated to >>50°C.
Heat is generated when charged ions in tissue oscillate rapidly (following the alternating RF current) converting RF energy to kinetic/thermal energy (Ohmic/Resistive Heating)
3/10
According to Ohm’s law, the amount of power per unit volume (resistive heating) equals the square of current density times the impedance (resistance) of the tissue, which in turn, is a function of the square of RF current density.
4/10
RF current density (and power dissipation per unit volume) decreases rapidly (with the 4th power of distance) as it distributes radially from abl electrode. Thus, only a thin (<2 mm) rim of tissue in contact with the electrode is directly heated by RF (Resistive Heating)
5/10
Heat is then conducted into deeper layers of myocardium (Conductive Heating), which extends the abl lesion size.
Heat is also conducted back to the circulating blood pool and to the metal ablation electrode. This heat loss opposes myocardial heating and lesion formation.
6/10
When power output is fixed, a larger abl electrode creates smaller lesion size by 2 mechanisms: (1) It exposes more of the electrode surface area to the cooling effect of blood. (2) It reduces RF current density (due to larger surface area), which reduces tissue heating.
7/10
Active cooling of the abl electrode and tissue surface by irrigation increases the amount of heat loss from the tissue (acting as a heat sink) and can reduce lesion size in the setting of a fixed RF power delivery.
8/10
In our question, power delivered to tissue is fixed. However, the proportion of heat stolen from tissue by abl electrode & blood is least (& current density is higher) with small non-irrigated electrode. Thus, heat remaining in tissue is maximized & lesion size is larger.
9/10
So, why do we use large or irrigated abl electrodes?
The temperature of small non-irrigated abl electrodes rises quickly, which limits RF energy delivery (in temp-controlled mode) and can precipitate coagulum formation, which increases impedance and limits power delivery.
10/10
When RF power output is unlimited, cooling of the abl electrode passively (by using a larger electrode size) or actively (by using saline irrigation) helps to avoid electrode-tissue interface overheating, allowing for greater power delivery and creation of larger lesions.
• • •
Missing some Tweet in this thread? You can try to
force a refresh
“Aberration” describes transient bundle branch block (BBB) and does not include persistent QRS abnormalities caused by persistent BBB, preexcitation, or the effect of drugs.
Acceleration-dependent BBB (aka “phase 3 block” or “voltage-dependent block”) occurs when an impulse arrives at tissues that are still refractory due to incomplete repolarization (during phase 3 of the action potential [AP]).
3/9
Aberration secondary to phase 3 block tends to be in the form of RBBB when premature excitation (and Ashman phenomenon) occurs during normal baseline heart rates and in the form of LBBB when it occurs during fast heart rates.
Impedance and RF ablation:
Part 2: How does RF ablation affect impedance?
1/9 As tissue temperature rises during RF energy application, ions within the tissue being heated become more mobile, resulting in a decrease in impedance to current flow.
2/9 There are currently 2 methods to measure impedance: Generator Impedance (GI) & Local Impedance (LI).
3/9 Lack of impedance drop during RF energy application can reflect inefficient energy delivery to the tissue due to poor tissue contact, lack of catheter stability, or inadequate power delivery.
Impedance & RF ablation:
Part 1: How does impedance affect RF lesion formation?
1/8 During RF ablation, system impedance = impedance of genera¬tor + transmission lines + catheter + electrode-tissue interface + skin patch interface + interposed tissues.
2/8 IMPEDANCE & POWER
The magnitude of RF current delivered by the generator is determined by impedance btwn ABL electrode and ground pad. Ablation at lower impedance yields higher current output (and tissue heating) compared with ablation at a similar power & higher impedance.
3/8 IMPEDANCE OF ELECTRICAL CONDUCTORS
Currently used electrical conductors from the generator to the patient and from the ground pad back to the generator are designed to have low electrical resistance to help minimize power loss within those conductors.
1/8 ECG patterns that mimic 2°AVB are often related to atrial ectopy, concealed junctional ectopy, or AVN echo beats. Distinguishing physiologic from pathologic AVB is important.
2/8 In 2°AVB, sinus P-P interval is fairly constant (except for some variation caused by ventriculophasic arrhythmia), the nonconducted P wave occurs on time as expected, and P wave morphology is constant. With ectopy, P waves occur prematurely & often have different morphology.
3/8 Early PACs can arrive at the AVN during the refractory period and conduct with long PRI or block (physiologic rather than pathologic block) and can mimic Mobitz I or Mobitz II 2°AVB.
#IssaTweetorials
What Is the Gap Phenomenon? 1/4 “PROXIMAL DELAY ALLOWS DISTAL RECOVERY” is the fundamental concept of gap phenomenon. This requires a distal site with a long effective refractory period (ERP) and a proximal site with a shorter ERP.
2/4 During gap phenomenon, initial block occurs distally (due to longer ERP). Earlier impulses encroach on the relative refractory period (RRP) of proximal site where conduction delay is encountered, which allows for expiration of the ERP of the distal site, enabling conduction.
3/4 Any pair of structures in the AV conduction system that has the appropriate EP physiological relationship can exhibit the gap phenomenon (e.g., atrium–AVN, proximal AVN–distal AVN, AVN–HPS, HB-distal HPS). Gap can occur in the anterograde or retrograde direction.
1/7 Concealed conduction can be defined as "the propagation of an impulse within the conduction system that can be recognized only from its effect on the subsequent impulse, interval, or cycle."
2/7 Impulse propagation in the conduction system generates too small electrical current to be recorded on ECG. If this impulse travels only a limited distance (incomplete penetration) in the conduction system, it can interfere with formation or propagation of another impulse.
3/7 Irregular Ventricular Response During AF:
AVN is expected to conduct at regular intervals when its RP expires after each conducted AF impulse. Irregular response is caused by incomplete penetration of some AF impulses into AVN, variably resetting its refractoriness.