Impedance & RF ablation:
Part 1: How does impedance affect RF lesion formation?
1/8 During RF ablation, system impedance = impedance of genera¬tor + transmission lines + catheter + electrode-tissue interface + skin patch interface + interposed tissues.
2/8 IMPEDANCE & POWER
The magnitude of RF current delivered by the generator is determined by impedance btwn ABL electrode and ground pad. Ablation at lower impedance yields higher current output (and tissue heating) compared with ablation at a similar power & higher impedance.
3/8 IMPEDANCE OF ELECTRICAL CONDUCTORS
Currently used electrical conductors from the generator to the patient and from the ground pad back to the generator are designed to have low electrical resistance to help minimize power loss within those conductors.
4/8 IMPEDANCE AT SKIN-GROUND PAD INTERFACE
Part of the RF power is dissipated at the skin-ground pad interface. High impedance at that location results in more RF power loss and skin heating at the ground pad, and less energy available for myocardial heating at ABL electrode.
5/8 IMPEDANCE AT SKIN-GROUND PAD INTERFACE
A large ground pad surface area (or adding a second patch) and meticulous skin preparation to optimize skin contact are required at to reduce impedance and minimize power loss, and to effectively dissipate heat and prevent skin burns.
6/8 IMPEDANCE AT ELECTRODE-BLOOD-TISSUE INTERFACE
With ABL electrode in contact with the endocardial wall, part of the electrode contacts tissue and the rest contacts blood, and the RF current flows through both myocardium & blood, then through the thorax to the ground pad.
7/8 RF LESION FORMATION
With normal electrode-tissue contact, only a fraction of all power is effectively applied to the tissue. The rest is dissipated in the blood pool and elsewhere in the patient.
8/8 MODULATING IMPEDANCE TO IMPROVE RF LESION FORMATION
RF current flow into the myocardium can be increased by reducing total circuit impedance and by modulating the ratios of electric impedances between RF electrode, blood pool, and cardiac tissue.
• • •
Missing some Tweet in this thread? You can try to
force a refresh
“Aberration” describes transient bundle branch block (BBB) and does not include persistent QRS abnormalities caused by persistent BBB, preexcitation, or the effect of drugs.
Acceleration-dependent BBB (aka “phase 3 block” or “voltage-dependent block”) occurs when an impulse arrives at tissues that are still refractory due to incomplete repolarization (during phase 3 of the action potential [AP]).
3/9
Aberration secondary to phase 3 block tends to be in the form of RBBB when premature excitation (and Ashman phenomenon) occurs during normal baseline heart rates and in the form of LBBB when it occurs during fast heart rates.
Impedance and RF ablation:
Part 2: How does RF ablation affect impedance?
1/9 As tissue temperature rises during RF energy application, ions within the tissue being heated become more mobile, resulting in a decrease in impedance to current flow.
2/9 There are currently 2 methods to measure impedance: Generator Impedance (GI) & Local Impedance (LI).
3/9 Lack of impedance drop during RF energy application can reflect inefficient energy delivery to the tissue due to poor tissue contact, lack of catheter stability, or inadequate power delivery.
1/8 ECG patterns that mimic 2°AVB are often related to atrial ectopy, concealed junctional ectopy, or AVN echo beats. Distinguishing physiologic from pathologic AVB is important.
2/8 In 2°AVB, sinus P-P interval is fairly constant (except for some variation caused by ventriculophasic arrhythmia), the nonconducted P wave occurs on time as expected, and P wave morphology is constant. With ectopy, P waves occur prematurely & often have different morphology.
3/8 Early PACs can arrive at the AVN during the refractory period and conduct with long PRI or block (physiologic rather than pathologic block) and can mimic Mobitz I or Mobitz II 2°AVB.
#IssaTweetorials
What Is the Gap Phenomenon? 1/4 “PROXIMAL DELAY ALLOWS DISTAL RECOVERY” is the fundamental concept of gap phenomenon. This requires a distal site with a long effective refractory period (ERP) and a proximal site with a shorter ERP.
2/4 During gap phenomenon, initial block occurs distally (due to longer ERP). Earlier impulses encroach on the relative refractory period (RRP) of proximal site where conduction delay is encountered, which allows for expiration of the ERP of the distal site, enabling conduction.
3/4 Any pair of structures in the AV conduction system that has the appropriate EP physiological relationship can exhibit the gap phenomenon (e.g., atrium–AVN, proximal AVN–distal AVN, AVN–HPS, HB-distal HPS). Gap can occur in the anterograde or retrograde direction.
1/7 Concealed conduction can be defined as "the propagation of an impulse within the conduction system that can be recognized only from its effect on the subsequent impulse, interval, or cycle."
2/7 Impulse propagation in the conduction system generates too small electrical current to be recorded on ECG. If this impulse travels only a limited distance (incomplete penetration) in the conduction system, it can interfere with formation or propagation of another impulse.
3/7 Irregular Ventricular Response During AF:
AVN is expected to conduct at regular intervals when its RP expires after each conducted AF impulse. Irregular response is caused by incomplete penetration of some AF impulses into AVN, variably resetting its refractoriness.
1/9 How reliable is ablation electrode temperature to monitor RF lesion formation?
RF lesion creation is thermally mediated; hence, monitoring tissue temp is the most reliable approach to achieve effective (temp >50°C) & safe (temp <100°C) ablation.
#EPeeps
2/9 Temp measurements relevant to RF lesion formation include:
Directly measuring intramyocardial temp is the optimal method to monitor RF abl, but such technologies are currently not available for clinical use. Investigational technologies: microwave radiometery, US thermal strain imaging, and MR thermometry.