Ziad F. Issa Profile picture
Oct 29, 2021 8 tweets 4 min read Read on X
#IssaTweetorials
#EPeeps

Impedance & RF ablation:
Part 1: How does impedance affect RF lesion formation?

1/8
During RF ablation, system impedance = impedance of genera¬tor + transmission lines + catheter + electrode-tissue interface + skin patch interface + interposed tissues.
2/8
IMPEDANCE & POWER
The magnitude of RF current delivered by the generator is determined by impedance btwn ABL electrode and ground pad. Ablation at lower impedance yields higher current output (and tissue heating) compared with ablation at a similar power & higher impedance.
3/8
IMPEDANCE OF ELECTRICAL CONDUCTORS
Currently used electrical conductors from the generator to the patient and from the ground pad back to the generator are designed to have low electrical resistance to help minimize power loss within those conductors.
4/8
IMPEDANCE AT SKIN-GROUND PAD INTERFACE
Part of the RF power is dissipated at the skin-ground pad interface. High impedance at that location results in more RF power loss and skin heating at the ground pad, and less energy available for myocardial heating at ABL electrode.
5/8
IMPEDANCE AT SKIN-GROUND PAD INTERFACE
A large ground pad surface area (or adding a second patch) and meticulous skin preparation to optimize skin contact are required at to reduce impedance and minimize power loss, and to effectively dissipate heat and prevent skin burns.
6/8
IMPEDANCE AT ELECTRODE-BLOOD-TISSUE INTERFACE
With ABL electrode in contact with the endocardial wall, part of the electrode contacts tissue and the rest contacts blood, and the RF current flows through both myocardium & blood, then through the thorax to the ground pad.
7/8
RF LESION FORMATION
With normal electrode-tissue contact, only a fraction of all power is effectively applied to the tissue. The rest is dissipated in the blood pool and elsewhere in the patient.
8/8
MODULATING IMPEDANCE TO IMPROVE RF LESION FORMATION
RF current flow into the myocardium can be increased by reducing total circuit impedance and by modulating the ratios of electric impedances between RF electrode, blood pool, and cardiac tissue.

• • •

Missing some Tweet in this thread? You can try to force a refresh
 

Keep Current with Ziad F. Issa

Ziad F. Issa Profile picture

Stay in touch and get notified when new unrolls are available from this author!

Read all threads

This Thread may be Removed Anytime!

PDF

Twitter may remove this content at anytime! Save it as PDF for later use!

Try unrolling a thread yourself!

how to unroll video
  1. Follow @ThreadReaderApp to mention us!

  2. From a Twitter thread mention us with a keyword "unroll"
@threadreaderapp unroll

Practice here first or read more on our help page!

More from @ZiadIssaMD

Dec 10, 2021
#IssaTweetorials
 
What is the mechanism of aberrant conduction?
 
1/9

“Aberration” describes transient bundle branch block (BBB) and does not include persistent QRS abnormalities caused by persistent BBB, preexcitation, or the effect of drugs.

#EPeeps #CardioTwitter #ECG
2/9

Acceleration-dependent BBB (aka “phase 3 block” or “voltage-dependent block”) occurs when an impulse arrives at tissues that are still refractory due to incomplete repolarization (during phase 3 of the action potential [AP]).
3/9

Aberration secondary to phase 3 block tends to be in the form of RBBB when premature excitation (and Ashman phenomenon) occurs during normal baseline heart rates and in the form of LBBB when it occurs during fast heart rates.
Read 9 tweets
Nov 17, 2021
#IssaTweetorials
#EPeeps

Impedance and RF ablation:
Part 2: How does RF ablation affect impedance?
 
1/9
As tissue temperature rises during RF energy application, ions within the tissue being heated become more mobile, resulting in a decrease in impedance to current flow. Image
2/9
There are currently 2 methods to measure impedance: Generator Impedance (GI) & Local Impedance (LI).
3/9
Lack of impedance drop during RF energy application can reflect inefficient energy delivery to the tissue due to poor tissue contact, lack of catheter stability, or inadequate power delivery.
Read 8 tweets
Oct 8, 2021
#IssaTweetorials

The mimicry of second-degree AV block (2°AVB)

1/8
ECG patterns that mimic 2°AVB are often related to atrial ectopy, concealed junctional ectopy, or AVN echo beats. Distinguishing physiologic from pathologic AVB is important.

#EPeeps #CardioTwitter #ECG
2/8
In 2°AVB, sinus P-P interval is fairly constant (except for some variation caused by ventriculophasic arrhythmia), the nonconducted P wave occurs on time as expected, and P wave morphology is constant. With ectopy, P waves occur prematurely & often have different morphology.
3/8
Early PACs can arrive at the AVN during the refractory period and conduct with long PRI or block (physiologic rather than pathologic block) and can mimic Mobitz I or Mobitz II 2°AVB.
Read 8 tweets
Oct 1, 2021
#IssaTweetorials
What Is the Gap Phenomenon?
1/4
“PROXIMAL DELAY ALLOWS DISTAL RECOVERY” is the fundamental concept of gap phenomenon. This requires a distal site with a long effective refractory period (ERP) and a proximal site with a shorter ERP.

#EPeeps #CardioTwitter #ECG Image
2/4
During gap phenomenon, initial block occurs distally (due to longer ERP). Earlier impulses encroach on the relative refractory period (RRP) of proximal site where conduction delay is encountered, which allows for expiration of the ERP of the distal site, enabling conduction. ImageImage
3/4
Any pair of structures in the AV conduction system that has the appropriate EP physiological relationship can exhibit the gap phenomenon (e.g., atrium–AVN, proximal AVN–distal AVN, AVN–HPS, HB-distal HPS). Gap can occur in the anterograde or retrograde direction. Image
Read 4 tweets
Sep 24, 2021
#IssaTweetorials
What is Concealed Conduction?

1/7
Concealed conduction can be defined as "the propagation of an impulse within the conduction system that can be recognized only from its effect on the subsequent impulse, interval, or cycle."

#EPeeps
#CardioTwitter
#ECG
2/7
Impulse propagation in the conduction system generates too small electrical current to be recorded on ECG. If this impulse travels only a limited distance (incomplete penetration) in the conduction system, it can interfere with formation or propagation of another impulse.
3/7
Irregular Ventricular Response During AF:
AVN is expected to conduct at regular intervals when its RP expires after each conducted AF impulse. Irregular response is caused by incomplete penetration of some AF impulses into AVN, variably resetting its refractoriness.
Read 7 tweets
Sep 17, 2021
#IssaTweetorials

1/9
How reliable is ablation electrode temperature to monitor RF lesion formation?

RF lesion creation is thermally mediated; hence, monitoring tissue temp is the most reliable approach to achieve effective (temp >50°C) & safe (temp <100°C) ablation.

#EPeeps Image
2/9
Temp measurements relevant to RF lesion formation include:

1- Intramural temp.
2- Endocardial surface temp.
3- Abl electrode temp.
3/9
Intramural Tissue Temp:

Directly measuring intramyocardial temp is the optimal method to monitor RF abl, but such technologies are currently not available for clinical use. Investigational technologies: microwave radiometery, US thermal strain imaging, and MR thermometry.
Read 8 tweets

Did Thread Reader help you today?

Support us! We are indie developers!


This site is made by just two indie developers on a laptop doing marketing, support and development! Read more about the story.

Become a Premium Member ($3/month or $30/year) and get exclusive features!

Become Premium

Don't want to be a Premium member but still want to support us?

Make a small donation by buying us coffee ($5) or help with server cost ($10)

Donate via Paypal

Or Donate anonymously using crypto!

Ethereum

0xfe58350B80634f60Fa6Dc149a72b4DFbc17D341E copy

Bitcoin

3ATGMxNzCUFzxpMCHL5sWSt4DVtS8UqXpi copy

Thank you for your support!

Follow Us!

:(