Caveat 1: earlier grid had an error, giving readings incorrect dispersion.
Caveat 2: Libre2 data from an earlier experiment due to device failing early in this test and replacement taking 10 days.
Thread.
/1
Calibration was done according to manufacturer directions:
Dexcom ONE & G6 - never;
Aidex - when I felt it was needed;
Glucomen Day - every 24 hours;
Medtrum Nano - Factory calibration was enabled, but calibration was done when it appeared to drift too much;
Libre2 - never.
/2
Dexcom ONE and Dexcom G6 have almost identical dispersion, as you might expect given the sensor and calibration technology is identical:
/3
GlucoRX Aidex has a wider dispersion than either of the Dexcom devices:
/4
This is the Glucomen Day:
/5
This is the Medtrum Nano:
/6
And finally, this is the Libre2, although using a different set of reference data:
/7
What's clear when looking at these results side-by-side is that the dispersion of points in relation to the reference data is much wider for the @GlucoRX Aidex, @GlucoMen Day and @medtrumdiabetes Nano than for #Libre2 or either Dexcom device.
/8
What's perhaps more concerning is that at "out of range" glucose levels (<80mg/dl or >180mg/dl), there appears to be far more inaccuracy in the three newcomers.
The question is, if you were about to drive, which of these sensors is reliably safe?
/9
And secondly, which could you safely use in an Open Source #WeAreNotWaiting automated insulin delivery device?
I wouldn't be comfortable with some of these.
/10
Further data to come over at diabettech.com, but that's the heads up, quick and dirty review.
If you're offered any of these on the NHS, which would you prefer?
So the new @Glucomen sensor has been attached. Here's how you configure the inserter: /1
And then put it into the insertion device... /2
Which then requires two button presses to apply the sensor then remove the inserter. It didn't hurt though, and I felt it slightly less than the G6 (which I did at the same time). /3