[1] Accurate germicidal efficacy and photobiological hazard measurement of far-UVC sources.
Mike Clark, Gigahertz Optik GmbH #ICFUST2023 #ICFUST
[2] Need better meters especially between fluence and irradiance. UV radiomerters - widely used, convenient
UVspectroradiometers - much more comprehensive
[3] Radiometers often calibrated for near radiation. Most commercial radiometers don't cover the range of Far UV-C. Also need to match spectral response to wavelength of interest or big errors. Knowledge of light source used in calibration is very important!
[4] In-situ hazard measurements are very low, so you need a lot of sensitivity, and complex because of spectral distribution. Out of band response can result in overestimate of risk.
[5] IEC 62471:2006 requires a double monochromator system or a system that can be correlated. Most people are using array spectrometers. However internal stray light performance is a limiting factor for UV array-spectroradiometers. Manf. quote these but need to know how measured
[6] Plot showing simulated spectrum error impacts on a ACGIH level type measurements [Can't we just use the 222nm as the reference itself]
[7] UV spectroradiometer requires calibration with a duterium lamp, if using integrating sphere watch out for fluorescence!
[8] Conclusion - Need good cosine response to get total irradiance
Radiometers - Need spectral response to match light of interest
Spectroradiometers - Need calibration light source, and accuracy of wavelength important.
• • •
Missing some Tweet in this thread? You can try to
force a refresh
[2]written in direct response to the conference (and during the conference) reminder: COVID has caused enormous caseload and death and pre COVID lower respiratory infection was already #4 cause of worldwide death.
[3] This is a huge health challenge worldwide
three vectors for air and infection: communicable diseases( TB measles, flu), opportunistic pathogens (aspergillusm, NTMs, etc), airborne role in other infections (mrsa, c diff, norovirus)
[1]How it goes forward: A blueprint for communication
with the public about GUV
Kathleen McPhaul, University of Maryland #ICFUST #ICFUST2023
[2] (First shout to Balvi for the support on this one! (
Going to address health literacy. The context is the extreme importance of getting the messaging right and letting this space make the change it can
[3] Examples of health messaging that went off track from a public health messaging perspective:
mask effectiveness: they work, they worked, that message was missed. Vaccinations work, but the term "break through" diminished (as opposed to "saving lives, decreasing illness, etc"
[1] Roundtable discussion with participants from both governmental and non-governmental agencies on future regulatory pathways for Far-UVC #ICFUST #ICFUST2023
[2] Moderator: Lew Radonovich (CDC, NIST)
Participants: Belal Abboushi (Pacific Northwest National Laboratory), Aline Heffernan (EPA Office of Pesticide Programs), Cameron Miller (NIST), David Sliney (perspectives from ACGIH), Richard Vincent (perspectives from ASHRAE)
[3] Lew Radonovich (CDC) opens that we will discuss the regulatory space(or really lack of)
NIOSH in US is focused on occupational safety and health, seek knowledge and apply it to the workplace. Will move through the panel now.
[1] Germicidal Ultraviolet Light dosing
recommendations from the field
PJ Piper, Far UV Technologies
[2] Overview of Upper-Room vs Whole Room(222 nm). Aside: Airplanes hard, need $300M insurance policy so not happening any time soon.
[3] Club Cafe - Boston, MA example install.
Floorplan important, and ceiling heights important.
Originally discussed with clients log kill rate, but will transition to eACH how adjacent industry talk
[1] Goniophotometric intensity and spectral measurements in the UV for a Care222 Excimer lamp unit with and without a filter and diffuser unit
Rolf S. Bergman, Rolf Bergman Consulting
[2] We care about total photonic output for disinfection. Wanted to explore spectrum impacts off axis, and what happens with difuser.
Lamps: USHIO B1 with and with out filter, and also with added diffuser
[3] Data shows angular difference with and without filters in terms of 222nm output(as expected). Spectral data with off-axis shows reduction in output as well(expected) and the filter reduced light above 235nm which is good.
[1] Impact of test methods on Far-UVC performance against bioaerosols
Katherine Ratliff, U.S. Environmental Protection Agency #ICFUST #ICFUST2023
[2] How effect? how to compare tech types? How can lab results be used to inform predictions?
How do extrapolate lab tests to real world applications?
[3] Use large bioaerosol test chamber(10x12x25 feet) with mock HVAC system.
Use MS2 as surrogate, and most importantly also did replicates.
Quantify log reduction and CADR of 222nm lamps and portable HEPA air cleaners.