AV Rails are useful in complex paravalvular leak closure, but not commonly discussed in the literature
This short🧵sheds some light on AV rails & their attributes.
#CardioTwitter #MedEd
💢When should I I use a rail?
📌 Simple leaks don’t require a rail
📌 Rail upfront in serpiginous/Ca++ leaks. I use it in 1/3 of cases. When in doubt, use a rail!
📌 Rails are also great opportunity to ‘electively’ master snaring 😉
But sometimes rails can be challenging👇
💢 e.g. Presence of 2 mechanical valves. Here u have 3 options:
1. Transapical rail
Caveat: TA🩸risk
2. AV rail across the mechanical AV
Caveat: leaflet impingement (can often be done carefully)
3. VV rail if double MV leaks: (aka LAMPOON style)
Caveat: valve instability
💢 Can also use other sources of support as an alternative to rail
In👇case of 2 mechanical valves and 2 MV leaks
- Planned to make a VV rail through the leaks (like a LAMPOON)
- But, skipped the rail as leaning the Agilis on the RA free wall providing adequate back support
💢 Aortic PVL can also be challenging & railing may help
In👇case:
- Plugs couldn’t advance retrograde but went through antegrade with an AV rail
- Apical rail considered but avoided due to prior TA TAVR
- Caveat: need to carefully watch the IAS to avoid tears or large ASD
💢 Final thoughts:
- Longest wires possible:
Glide ES, Nitrex 400 cm
- EnSnares >> Gooseneck
- Keep 👁 on AO: AV rails cause AI which can be ⬇️ by easing the rail
- Like with anchor wire techniques, the rail wire will occupy space, consider👇 when selecting u plug