- A Tweetorial
⚡️What is a ‘Stuck’ HD catheter?
⚡️What are the risk factors?
⚡️How to manage it?
@ASDINNews
#VascualarAccessPearls
1/
⚡️Commonly referred to as a ‘Hemodialysis catheter that won’t come out’
But there is more to it...
2/
⚡️It likely occurs due to fibrin sheath formation around the CVC & subsequent adherence of the sheath to the vessel wall
3/
4/
⚡️If the CVC remains stuck despite cuff dissection then cutdown over the CVC near the neck can be done
5/
If the HD catheter remains stuck despite cuff dissection & cutdown then the next best step to remove the stuck HD catheter is:
6/
7/
⚡️Severe pain
⚡️Avulsion of the great vessels
⚡️Damage to the tricuspid valve
⚡️Catheter fracture
8/
⚡️It is rare & the incidence is ~1%
👉🏽pubmed.ncbi.nlm.nih.gov/25953209/
⚡️But because CVC use is very common, you are very likely to encounter this problem in your clinical practice
9/
⚡️Cumulative duration of CVC use (>1 year) appears to ⬆️ the risk of catheter adhesion
⚡️This is why some have suggested annual prophylactic CVC exchange (‼️This is not done in practice)
10/
Other potential risk factors include:
⚡️Female gender
⚡️Composition of the CVC
⚡️LIJ CVC
⚡️Prior CVC stenosis
11/
⚡️In 2011, Hong et al described the endoluminal ballon dilatation technique to remove a stuck HD CVC
👉🏽 pubmed.ncbi.nlm.nih.gov/21688240/
⚡️Subsequent reports have confirmed success with this technique
12/
peri-catheter adhesions facilitating the catheter removal
14/
👉🏽 ncbi.nlm.nih.gov/pmc/articles/P…
15/
⚡️Removal of the CVC hub & use of a vascular sheath around the CVC for extraction
⚡️Use of snares for extraction
⚡️Median Sternotomy & Thoracotomy
16/
⚡️Infection
⚡️Thromboembolism
⚡️Chronic pain
17/
If the HD catheter remains stuck despite cuff dissection & cutdown, then the next best step to remove the stuck HD catheter is:
18/
⚡️Stuck HD catheters are rare but can be seen w/ prolonged CVC use
⚡️If CVC cuff dissection & cutdown fail in removing the HD catheter then endoluminal balloon dilatation is the preferred intervention for removal of a ‘stuck’ HD catheter
End/