What's going on with the Impella?

As a follow up to my previous #tweetorial on the IABP, this🧵 will explore the nuts and bolts of the Impella for those in the CCU

Special thanks to @AHajduczok for helping with this🧵.
#medtwitter #cardiotwitter
@AHajduczok Learning objectives:
💔 history of the Impella
💔 learn how it works
💔 how to troubleshoot when it doesn't work
@AHajduczok The Impella is based on concept from Egypt in the 3rd Century BC

The Greek mathematician Archimedes, developed a machine to raise water up while in Egypt. The machine consists of a screw within a hollow cylinder that transfers water upwards by each segment of the screw.
@AHajduczok Dr. Richard Wampler adapted the design in 1985 by inventing the Hemopump, the first intravascular axial flow LV support device.

In 1988, the Hemopump was successfully implanted in a patient at @Texas_Heart. However, it was never a commercial success and was soon discontinued.
@AHajduczok @Texas_Heart In 1991, Dr. Thorsten Siess modified the Hemopump to include a shorter rotor and the addition of a motor on the catheter itself.

This gave rise to the Impella class of products that were approved for clinical use in Europe (2005) and the US (2008).

ncbi.nlm.nih.gov/pmc/articles/P…
@AHajduczok @Texas_Heart What are the important components of the Impella?

Inlet: 4 (or 5) openings that draws blood into cannula

Cannula: spiral shaped body made of nitinol and polyurethane

Outlet: where blood exits cannula

Motor: powers rotor in outlet

Marker: for positioning (only on 2.5 and CP)
@AHajduczok @Texas_Heart Indications include:
💔 cardiogenic shock immediately (<48 hours) following acute MI or CABG

💔 cardiogenic shock in the setting of cardiomyopathy or myocarditis

💔 during high risk PCI
@AHajduczok @Texas_Heart Contraindications include:
🚫 LV pathology (thrombus, rupture, VSD)

🚫 Aortic valve pathology (Mod/severe AR, significant AS, mechanical aortic valve)

🚫 Significant right ❤️ failure (can also use RP for RV support)

🚫 Cardiac tamponade

🚫 Severe PAD that precludes placement
@AHajduczok @Texas_Heart Duration of use is dependent on the model and indication:

For cardiogenic shock:
💓<4 days for Impella 2.5 and CP
💓<14 days for Impella 5.0, 5.5, LD, and RP

For high risk PCI:
💓 <6 hours for Impella 2.5 and CP
@AHajduczok @Texas_Heart What's the difference between the models?

⚡ The numbers indicate the maximum hemodynamic support by the model

⚡ The later models (RP, 5.0, LD, and 5.5) all feature a differential pressure sensor

⚡ All models except for 2.5 feature Smart Assist®️
@AHajduczok @Texas_Heart What is the differential pressure sensor?

It's a membrane that flexes when there is a pressure differential between the outer surface (aortic pressure) and inner surface (ventricular pressure)

This is an important concept related to the placement signals on the monitor
@AHajduczok @Texas_Heart What is Smart Assist®️ ?

It's an optical sensor that assists with catheter repositioning without image guidance

It also provides more advanced hemodynamic data such as:
⚡ LVEDP
⚡ MAP
⚡ cardiac power output (CPO) = [(MAPxCO)/451]
@AHajduczok @Texas_Heart What information is found on the Impella monitor?

Placement signal: pressure differential between LV and aorta (⬆️ in diastole, ⬇️ in systole)

Motor current: a measure of load on the device (⬆️ in systole, ⬇️ in diastole)

Flow control: where you would find P-level (more next)
@AHajduczok @Texas_Heart What's up with the Performance (P) level?

It is a measure of flow that corresponds to both motor current and RPMs of the device

Factors such as improper positioning and suction events can reduce flow rates at a given P level

...but what is the right position??
@AHajduczok @Texas_Heart Using both TTE and TEE, optimal positioning includes (see 👇):
1️⃣ inlet 3.5 cm below aortic valve
2️⃣ outlet well above the aortic valve
3️⃣ catheter angled toward the LV apex
@AHajduczok @Texas_Heart Correct positioning can also be confirmed through color doppler.

If the catheter is positioned properly, you would expect turbulent flow above the aortic valve (indicating the outlet is above the valve)
@AHajduczok @Texas_Heart Now let's cover some of the common issues that can arise with the Impella
💔 improper positioning
💔 Suction alarms
💔 Hemolysis
@AHajduczok @Texas_Heart Malpositioning can be identified by the waveforms

💔 if placed too far in the LV, you will see a ventricular waveform with a flattened motor current (A)

💔 if placed too high in aorta, you will see an aortic waveform with a flattened motor current (B)
@AHajduczok @Texas_Heart How do you manage when your Impella is malpositioned?

⬇️ P level to P2 and (normally under fluoroscopic or echo guidance) slowly advance/retract the catheter until the inlet is located 3.5cm below the aortic valve annulus

THEN, slowly re-establish flow up to the desired P level
@AHajduczok @Texas_Heart Suction alarms are another issue that can arise with your Impella

Causes include:
💔 malpositioning (notice a theme here)
💔 hypovolemia
💔 RV failure
💔 Pericardial effusion → tamponade

Here's a great summary of causes and ways to correct it below: cjcopen.ca/article/S2589-…
@AHajduczok @Texas_Heart Hemolysis 🩸 is the final not-so-subtle complication of the Impella, but how does it happen?

🩸 when blood travels through small spaces, velocity goes ⬆️ which ⬆️ shear forces → hemolysis

🩸 hemolysis also occurs when the catheter is malpositioned due to turbulent blood flow
@AHajduczok @Texas_Heart How can we recognize hemolysis?

First,
🩸 look for blood in the foley (see photo 👇)
🩸 is there a new decrease in CVP and PCWP?
🩸 is the Impella in the right position?

Lab findings
⬆️ LDH, indirect bilirubin, and plasma-free hemoglobin (see graph 👇 )
⬇️ Hgb, haptoglobin
@AHajduczok @Texas_Heart Once we identify hemolysis, how can we fix it?

🩸 Correct any positioning or suction alarms
🩸 repeat labs to prove that hemolysis is occurring
🩸 give back fluid or blood if CVP or PCWP <10mmHg
@AHajduczok @Texas_Heart Remember, do not touch the Impella if you are not familiar and trained with the device and only if and when you have approval from your attending. If you are a resident, call your fellow and/or attending immediately if these issues arise.
@AHajduczok @Texas_Heart Hopefully you found this🧵 helpful, and it will make your time in the CCU just a little bit easier!

Recommendations for further reading:
academic.oup.com/eurheartjsupp/…
pubmed.ncbi.nlm.nih.gov/35184086/

• • •

Missing some Tweet in this thread? You can try to force a refresh
 

Keep Current with Matt DiMeglio DO, MBA

Matt DiMeglio DO, MBA Profile picture

Stay in touch and get notified when new unrolls are available from this author!

Read all threads

This Thread may be Removed Anytime!

PDF

Twitter may remove this content at anytime! Save it as PDF for later use!

Try unrolling a thread yourself!

how to unroll video
  1. Follow @ThreadReaderApp to mention us!

  2. From a Twitter thread mention us with a keyword "unroll"
@threadreaderapp unroll

Practice here first or read more on our help page!

More from @Matt_DiMeglio

Apr 6
Can heart transplant recipients experience vasovagal syncope??

This came up during my time in the CCU, here's a short 🧵 investigating this question. Special thanks to @AHajduczok of @TJHeartFellows.

Let's first start with how vasovagal syncope normally occurs...
It starts with a trigger (stress, pain) coupled with central hypovolemia (dehydration or upright posture) → ⬆️ contractility of an underfilled LV

This ultimately results in ↑vagal efferent activity to sinus node (↓HR) and ↓sympathetic efferent activity to arterioles (↓SVR)
Remember that cardiac output = stroke volume x HR

The vasovagal reflex works synergistically to result in a profound ↓ in cardiac output that leads to ↓ cerebral blood flow → loss of consciousness
Read 12 tweets
Feb 21
What's going on with the intra-aortic balloon pump (IABP)?

Here's a practical🧵 for all residents and fellows during their time on rotation in the cardiac ICU.

Learning objectives
🎈 basics of the IABP
🎈 IABP alarms
🎈 ways to fix them

#CardioTwitter #Medtwitter #tweetorial
Let's start with the basics.

Developed in the 1950s by the Kantrowitz brothers, IABPs have been in clinical practice since the first implant in 1967 at @MaimonidesMC Currently, it is still widely used as modest temporary mechanical circulatory support for cardiogenic shock
The mechanism of IABP begins with rapid deflation of the helium filled balloon (helium moves faster than air) at the onset of systole. This creates a vacuum effect which reduces the work load of the heart during all of systole thereby reducing afterload
Read 28 tweets

Did Thread Reader help you today?

Support us! We are indie developers!


This site is made by just two indie developers on a laptop doing marketing, support and development! Read more about the story.

Become a Premium Member ($3/month or $30/year) and get exclusive features!

Become Premium

Don't want to be a Premium member but still want to support us?

Make a small donation by buying us coffee ($5) or help with server cost ($10)

Donate via Paypal

Or Donate anonymously using crypto!

Ethereum

0xfe58350B80634f60Fa6Dc149a72b4DFbc17D341E copy

Bitcoin

3ATGMxNzCUFzxpMCHL5sWSt4DVtS8UqXpi copy

Thank you for your support!

Follow Us on Twitter!

:(