[1] #Hemodynamics Tweetorial #2

Heart failure pt in ICU is -3.5L after 2d of aggressive diuresis. On day 3, urine output is ⬇️and BUN/Cr is ⬆️

You personally wedge #PAC at bedside and obtain a mean wedge 17 mmHg (a normal mean wedge is 6-10 mmHg). Admission wedge was 24
[2]
Question 1. Based on wedge 17 mm Hg (good waveform; no trick), you:
[3]
Before you use the pulmonary artery catheter #PAC information, you should know its pitfalls

Given the potential impact of #PAC-related decisions, consider learning as much (or more) as the ICU nurse/cath lab tech about #PAC setup and troubleshooting
[4]
🔹”Zeroing”
🔹”Leveling”
🔹”Damping”
[5]
🔹”Zeroing”

-Device is “zeroed” to open the air-fluid interface to atmospheric pressure (AP)

-AP is the reference standard against which other pressures are measured

-“Re-zeroing” is done often since transducer and atmospheric pressure drift away from calibration point
[6]
Transducer components and directions of the components 👇🏻
[7]
To zero:

-Point the 3-way stopcock “off” towards the patient (points up on the transducer)

-This now allows the vent port to be “opened” to atmospheric pressure

-If vent port cap is not fenestrated, it has to be temporarily removed
(port cap in video 👇🏻 is fenestrated)
[8]
🔹”Leveling”

Question 2. For #PAC, you place the transducer to match the level of:
[9]
🔹”Leveling”

-Selection of a point of interest (R atrium for #PAC) at which the ref standard (zero) is set

-RA is considered the phlebostatic axis (RA is also used as the ref point for CVP assessm)

*REMEMBER:
Pressure changes 7.4mmHg for every 10 cm transd is ⬆️ or ⬇️ RA
[10]

To level:
Identify the point at which the:
🔹mid-axillary line intersects with the
🔹4th intercostal space

That point should be at the same level of the transducer’s diaphragm

Consider using an accurate leveling system (e.g., laser) to match transducer ht with RA ht
[11]
Leveling is 🔑

🔷IN RELATIONSHIP TO TRANSDUCER:

🔹For every 10cm that RA “rises” above transducer, pressure falsely “rises” 7.4 mmHg

🔹For every 10cm that RA is “lowered” below transducer, pressure falsely “lowers” 7.4 mmHg



RA ⬆️ = pressure ⬆️

RA ⬇️= pressure ⬇️
[12]
Answer to Question 1 is: “Can I trust that wedge?”

In our case, when wedge 17 mmHg was obtained, transducer wasn’t leveled and pt’s RA was about 11 cm above the transducer.

Therefore, real wedge was <10 mmHg.

Diuresis was held for a few hrs and restarted at home dose
BONUS:

If you see pressure tracings that become negative (below Zero), consider an unaddressed LEVELING issue before you interpret and use that #PAC info

See 👇🏻 RAP tracing: transducer was higher than RA level, causing sub-Zero values (RA level is ⬇️ = pressure is “falsely” ⬇️)
[13]
🔹”Damping”

Damping is the “shock absorber” effect on the pressure line (like a car suspension)

Overdamping:
Sluggish oscillations

Underdamping:
Oscillations are too pronounced

Causes of Over and Under-Damping 👇🏻
[14]
Which answer is FALSE regarding an OVERDAMPED waveform
[15]
🔹OVER-damping

🔸Slurred upstroke
🔸ABSENT Dicrotic Notch
🔸Systolic is Underestimated
🔸Diastolic is Overestimated
🔸MAP is usually preserved
🔸Pulse pressure is underestimated (affects PAPi calc)
🔸The waveform gets “squished” (think letter “O” for “O”verdamped)
[16]

How to address an OVERDAMPED #PAC (and arterial) waveform
👇🏻
[17]
Which answer is FALSE regarding an UNDERDAMPED waveform:
[18]
🔹UNDER-damping

🔸Non-physiologic oscillations during Diast phase
🔸Deep dicrotic notch
🔸Systolic is Overestimated
🔸Diastolic is Underestimated
🔸Pulse pressure overestimation
🔸MAP is usually preserved
🔸Waveform gets “stretched out” (inverted “U” for “U”nderdamped”
[19]

How to address an UNDERDAMPED #PAC waveform
👇🏻
[20]

Besides looking at the waveforms, consider FAST-FLUSH test (square wave test) to assess for waveform damping characteristics

🔸Hold fast-flush device for <1 sec
🔸Assess the square wave and characteristics of the wave (s) that follow

See image 👇🏻
[21]

TAKE-HOME Messages (applies to ICU and outpt cardiac/PH pts)

🔷Before you interpret/use #PAC info, verify:
🔸Zeroing (atmosp pressure)

🔸Leveling (match transducer level to RA, AKA phlebostatic axis)

🔸Waveforms (to exclude OVER or UNDER damping affecting actual values)
TAKE-HOME Messages 2

🔸If you see -ve (below Zero) pressures, especially in non-ill pts, consider leveling error

🔸Over and Underdamping affect Syst and Diast pressures; MAP is usually preserved

🔸Verify line flushing/waveform every time a MVsat is drawn from distal #PAC port
Additional relevant reading:
-NEJM Art Line Pressure Transducer
nejm.org/doi/full/10.10…

-Deranged physiology
derangedphysiology.com/main/cicm-prim…

• • •

Missing some Tweet in this thread? You can try to force a refresh
 

Keep Current with Francisco J Soto, MD, MS

Francisco J Soto, MD, MS Profile picture

Stay in touch and get notified when new unrolls are available from this author!

Read all threads

This Thread may be Removed Anytime!

PDF

Twitter may remove this content at anytime! Save it as PDF for later use!

Try unrolling a thread yourself!

how to unroll video
  1. Follow @ThreadReaderApp to mention us!

  2. From a Twitter thread mention us with a keyword "unroll"
@threadreaderapp unroll

Practice here first or read more on our help page!

Did Thread Reader help you today?

Support us! We are indie developers!


This site is made by just two indie developers on a laptop doing marketing, support and development! Read more about the story.

Become a Premium Member ($3/month or $30/year) and get exclusive features!

Become Premium

Don't want to be a Premium member but still want to support us?

Make a small donation by buying us coffee ($5) or help with server cost ($10)

Donate via Paypal

Or Donate anonymously using crypto!

Ethereum

0xfe58350B80634f60Fa6Dc149a72b4DFbc17D341E copy

Bitcoin

3ATGMxNzCUFzxpMCHL5sWSt4DVtS8UqXpi copy

Thank you for your support!

Follow Us on Twitter!

:(