8/ How is effluent formed in CRRT?
It will depend in the modality used
✔️CVVH = hemofiltration
➡️🩸🩸🩸➡️
➖➖⬇️➖➖
E ⬅️💧💧
✔️CVVHD= Hemodialysis
➡️🩸🩸🩸➡️
➖➖➖➖
E ⬅️💧💧💧⬅️
✔️CVVHDF= Hemodiafiltration
➡️🩸🩸🩸➡️
➖➖⬇️➖➖
E ⬅️💧💧💧⬅️
9/Check how effluent is formed when using prismaflex with all the options activated
Remember: when giving convection every ml should be replaced with fluid in PRE filter, POST filter or both, so that fluid balance is 0
GIF 👇
CVVHDF
PRE + POST + Dialysate + fluid removal
10/ We can give the same dose and change the way the effluent is produced, only by changing modality IMAGE👇
🧐 If modality does not change the dose, why bother with different modality prescriptions?
11/ Three main reasons (in this order)
1. CRRT machine available 🎛
2. Technical aspects for circuit patency⚙️🔩
3. Type of transport needed (Convective or diffusive)
Lets review rationale for each one 🧐
12/ CRRT machines
⚡️The device available will be the most important variable on deciding a CRRT modality ⚡️
🧐Note: not necessarily the machine with more options is a better machine, there are other technical features that are not represented in the schema (other tweetorial)
13/ Circuit patency
⚡️Being an Intensity dependent therapy, circuit life should be the priority when choosing a modality⚡️
i.e.
-Using PRE RF to reduce FF will need a CVVH or CVVHD
-Using citrate as PRE dilution RF will need a CVVHDF
Until now we‘ve been talking about small size mol clearance(K)
But 🧐
-Medium size don’t saturate (E) to a 100% so (K) is ⬇️
-Convection will ⬆️ (K) of medium size mol by 30%~
-Small mol (K) will not change regardless of transport 👇
15/ Theoretically clearing middle (not measurable) “toxins” could improve outcomes & therefore using convection should do the job
🧐To this day there is no evidence that CONVECTION improves outcomes in the CRRT/AKI scenario & using it could lower ⬇️ circuit life
16/ To this point we know:
-(E) is used to dose CRRT because it represents (K) of small size molecules
-(E) forms depending on modality
-Modality choice depends on device availability, circuit life & type of transport
The next question will be
What dose?
17/ The recommended dose is 20-25 ml/kg/hr
This recommendation comes from a series of studies comparing high dose 40~ vs 20~, where no difference was found in mortality or adverse events
18/why not compare 20 vs lower dose? The reason is simple
CAVH studies demonstrated that 30~ ml/min urea (K) was needed to achieve metabolic control in catabolic well fed patients
✅20 ml/kg/hr is an EFFICIENCY metric that guarantees metabolic control to MOST patients (not all)
19/ Effluent ml/kg/hr is how CRRT is normally dosed, but are there other ways to do it ?
Absolutely, there are two interesting proposals that are based in urea clearance or urea kinetics that are worthy of mention 👇
20/ The first Proposed by @RClaure_nefro
Studied 6 Adequacy parameters
KB, KD, KtVB, KtVD, EKR, SRI & found that:
(K)(dialysis side) ml/min & Equivalent renal urea (K) were useful metrics to monitor & compare between dif treatments
✅(E) is used to dose CRRT because it represents (K) of small size molecules
✅(E)forms depending on modality
✅Modality choice depends on device availability, circuit life & type of transport
✅CONVECTION clears middle size mol better than DIFFUSION
22/ conclusions
✅To this day No evidence of better outcomes with CONVECTION
✅20ml/kg/hr is an EFFICACY metric that guarantees metabolic control in catabolic well fed patients
✅No evidence for higher dose
✅Other efficacy metrics proposed are KD & Std daily KtV
• • •
Missing some Tweet in this thread? You can try to
force a refresh
2/8👨🏽💼Male 50 years old needs CRRT for volume control and AKI
- 100 kg, 170 cm height, Htct 35%
- Modality CVVHDF
- Prescribed dose 25 ml/kg/hr
- Anticoagulation: none
- Institution down time average: 2 hours
- Urea cl dose based, so we use SC of 1
3/8The calculator will give you the amount of volume “dose” to distribute:
25x100kg——>2500 ml/hr
Distribute the volume at your preference
Dialysate —————1100 ml/hr
Replacement PRE —1100 ml/hr
Replecement POST -200ml/hr
Fluid Removal ———100 ml/hr