Also, the ventilator *might* be lying to you about its mode... #SoMe4MV
More on flow starvation here coemv.ca/flow-starvatio…
Given the robust inspiratory flow, the machine switched into *pressure support*! See the Servo manual for details. Compliments to @Thind888 on the ref
The canonical reference is here (in #FOAMcc form)
doi.org/10.4187/respca…
VC-
IMV (!)
d,d (dual, dual)
"Dual" means the mode of ventilation can change *within* a breath! And that is what is happening here with the "hump" in the waveform
As an example, what is PRVC actually?
This is
PC
CMV
a (adaptive)
The machine adjusts pressure (based on Pmusc) to reach a target Vt. There are major implications to this!
- Know the control variable, sequence, and targeting scheme for your ventilator(s)! It may be different than what you think
- Please don't leave a patient in flow starvation! Decide if you can liberalize their settings, or if you need tight control, take over!