How about a #tweetorial on acute pulmonary embolism (#PE) and how to use mechanical #thombectomy with @InariMedical #FlowTriever?? Lets goooooo!
#InariFellowsEdition #IRad #IC #cardiotwitter
Pic credit @jonathan_paulmd

📊100,000–180,000 PE-related deaths annually in US
📊PE is the most preventable cause of death among hospitalized patients
📊3rd leading cardiovascular cause of death after coronary disease and stroke
⏱ 1 American dies of a blood clot every 6 seconds


👍🏽GOOD at accommodating change in volume
👎🏽NOT GOOD at accommodating ACUTE changes in pulmonary pressure
#Cardiotwitter #InariFellowsForum #VTE #Hemodynamics

Patients with RVD defined as RV/LV >0.9 have a greater chance of adverse events within 30 days. Adverse event rate at 30 days:
☠️51.3% if RV/LV ratio ≤ 0.9
☠️80.3% if RV/LV ratio > 0.9
#InariFellowsForum



#echo #echocardiography #cardiotwitter #VTE @ASE360 @rajdoc2005 @DocStrom @echo_stepbystep @Ivan_Echocardio @ChamsiPash @hatemsoliman @PPibarot @Miroinmi @EchoAshwag #InariFellowsForum
2018 MGH #PERT Data Same As 20 Years Ago! Residual thrombus ➡️poor outcomes.
90 day mortality:
☹️41.3% for massive
☹️12.3% for intermediate (ICOPER rate 20 yrs ago: 14.7%)
2/3 of intermediate-risk deaths were post discharge


▪️FLAIR trial dem safety and ⬇️RV/LV
▪️20 F system. Powerful rapid aspiration of the clot (#Whoosh)
▪️Storage/retrieval of clot. 60 cc large bore syringe
Pic: @rishikumarmd -gotta love Cardiac Anesthesia!

🔑 Keys- let all them deploy and let them dwell! (Significant improvement ages 90 sec)
🔑 Crucial for removing chronic clot!
#VTE
✔️Large clot burden
✔️Submassive/massive PE
✔️#Lytic ineligible/actively bleeding
✔️Need for more urgent relief
✔️Able to tolerate large venous sheath insertion
✔️Able to tolerate 1-2 hour catheter procedure
#InariFellowsForum


🚨Special care must be taken given the size (20 Fr) of the FlowTriever System
Recommended:
🦢Swan or Baim Turi Catheter (PWP)- nice as allows 0.035 wire
▪️Alt: formed PIG
🖥Monitor hemodynamics and EKG
#InariFellowsForum


🐷Exchange for PIG and take PA pressure and angio
👤If pt can hold breath, perform DSA
🗣If unable to hold, cine angiography
🔖Use images to plan landing zone of device
🔗Two strategies for wire navigation: 0.035” or 0.018”
#InariFellowsForum
👀Pay attention to your wire!
🔄Loop is good!
❌Do not advance triever without the dilator!
🖇Consider buddy wire.
🔭Consider telescope technique with T16 inside T20
#InariFellowsForum #VTR @InariMedical @HadyLichaaMD @AntoniousAttall @DrAmirKaki
Perhaps the most important variable in max clot retrieval is: POSITION of T20 relative to the clot.
💝Optimal is 1-5 mm prox to clot for whoosh.
💞Consider “Double whoosh” if needed.
⚠️Keep wire and catheter position! ⏩Treat prox to distal.
#VTE