🧵What is GLS? 1/ It is evident that it is some measure of the systolic LV longitudinal shortening, normalised for the diastolic LV length, after the basic Lagrangian formula S = (L-L0)/L0
But how do we chose the numerator and denominator? 2/ The simplest measure would be LV systolic shortening / LV end diastolic length. In the HUNT 3 study, strain by this method was -17.1%. pubmed.ncbi.nlm.nih.gov/32978265/
LV shortening can be approximated by MAPSE, so GLS is similar to MAPSE normalised for LV diastolic length. Average MAPSE of sep-lat was similar to average of sep-ant-lat-inf within the measurement accuracy in the HUNT3 study. pubmed.ncbi.nlm.nih.gov/29399886/
But choosing the denominator is the tricky part. Chamber length, or wall length?
4/In the HUNT3 study an approximation to wall length, was the straight lines from mitral annulus to apex. Larger denominator: Normal strain by this method was - 16.3%.
But even that is not unambiguous. 5/ Do you draw the mid line or skewed lines to the endocardium, midway or epicardium?
And none of these use wall length, which is curved. 6/ Using the curved (true) wall length, would result in an even larger denominator, and a lower (absolute value) of GLS. Thus the closer we get to anatomy, the farther we move from the normal values by speckle tracking.
7/ Speckle tracking GLS shows normal values of -19% to - 21%. So why this discrepancy, speckle tracking should be incorporating wall curvature and follow the wall?
8/ Looking at a speckle tracking loop, what we see is that the tracking follows both longitudinal shortening and thickening of the wall.
9/ Inward motion of the wall is due to wall thickening, and is greatest in the endocardium, less in the midwall and least in the epicardium, because the innermost layers thicken into a smaller space, and thus thickens most. This inward motion will also shorten the wall lines:
10/ Thus speckle tracking strain is a composite of longitudinal and transmural (radial) strain, and over estimates the longitudinal strain per se, and is no gold standard. This would in itself not be a problem if the method was absolutely standardised, but it isn't.
11/ Firstly, inward tracking ability depend on speckle width, which again depends on a/ the scanner generation, b/ the chosen frame rate and c/ the image quality (near shadows result in smaller virtual aperture and reduced lateral resolution).
12/ as measured long. shortening is partly due to inward tracking, which increases towards the endocardium, longitudinal strain also depends on whether longitudinal strain is tracked a/ along the endocardium (e.g. VVI), b/ along the midline, or c/ as an average of the whole ROI
13/The shortening of longitudinal measurement due to inward motion depends on the curvature of the ROI, as illustrated by this example processed twice in the same loop
14/ -and we haven't touched into the technical details: How are speckles weighted according to a/ intensity, b/ stability, c/ drift and d/ amount of clutter filtering and c/ smoothing (usually AV plane motion is weighted highly, ant this motion is spline smoothed along the ROI)
15/ this means:
-no universal definitions
-geometrical differences
-user differences
-technical differences that to a large degree are industrial secrets, and thus vendor differences
Thus there is no gold standard, no possibility of validating GLS, and no universal normal limits
16/ My take home message is: GLS at its present level has no objective reference outside the system used, which means both scanner, application and user. Measures are not transferrable between vendors, scanner models, analysis software or even different generation of software.
17/ Also not from one user to another, and the user should make every effort do do analysis the same way every time. Standardisation of measurements may improve reproducibility, but there is no reference for comparison, and we do not know how it works across pathologies.
18/ EF has been the entry point in trials, but the weakness of EF is that it do not diagnose reduced systolic function where LVEDV and SV ar equally reduced. Long axis finction is much more sensitive, both for diagnosis and prognosis
19/ MAPSE, measures motion of the mitral ring. It only depends on M-mode, is generic and objectively verifiable, and is sensitive for reduced myocardial function, both in HFrEF and HFpEF, both in diag. and progn. Unfortunately, it's not sexy enough, so the GLS trials dominate.
• • •
Missing some Tweet in this thread? You can try to
force a refresh
However, in the HUNT study, we found no significant sex differences in MAPSE (although a trend, p=0.1), but in a large study of 1266 subjects, the difference was small < 0.05mm - far below measurement limit). pubmed.ncbi.nlm.nih.gov/29399886/ Why, when both are long axis function?
1/ In our study, we compared GLS derived from segmental values by our software, with MAPSE normalised for the corresponding end diastolic wall length (straight line) and non-normalised MAPSE pubmed.ncbi.nlm.nih.gov/29399886/
1/ During pre ejection, the vortex is seen to persist after MVC, and the septal part aligns with left ventricular outflow. This adds momentum and kinetic energy to the ejection flow.
2/ During ejection, however, the vortex seems to disappear, outflow more or less filling the whole apex, as flow in the lateral part is recruited by the rapid flow into the LVOT.
1/ The intraventricular vortex fills the ventricle, and the downwards flow in the septal part, will close the anterior MV leaflet. This also isolates the vortex in the ventricle, which may conserve the kinetic energy in the vortex
2/ At the end of diastasis, the lateral part of the vortex, with apical flow, is aligned with the incoming inflow in atrial systole, adding momentum and kinetic energy to the inflow during atrial systole.
🧵In our paper Intraventricular Vector Flow Imaging with Blood Speckle Tracking in Adults: Feasibility, Normal Physiology and Mech… we use a new method, not only BST, and can be applied on adult probes. pubmed.ncbi.nlm.nih.gov/34620522/
The main aim was to investigate the normal adult, intraventricular blood flow throughout the whole cardiac cycle, to compare with pw and colour Doppler M-mode and wall mechanics. (2D images courtesy of AS Daae).
As tweeted before, during IVR, there is simultaneous shortening of the base and elongation of the apex, inducing a volume shift with intraventricular apical flow, imparting a momentum and kinetic energy towards apex before start of early filling. This is thus *not* "wasted work"
Apart from the physiological implications, what are the consequences of this study onlinelibrary.wiley.com/doi/epdf/10.11… for timing of valve openings and closures by tissue Doppler?
1/ Valve closures can be timed by tissue Doppler and mitral ring motion. However, only the septal motion will reliably show AVC.
2/ MVO is close to the END of the pre ejection spike. Timing MVC by the start of the pre ejection spike will result in an error of about 40 ms too early. Timing by the peak R wave will result in about the same error.