Comparative Performance of SARS-CoV-2 Detection Assays using Seven Different Primer/Probe Sets and One Assay Kit
Spoiler alert: they all worked 😀 medrxiv.org/content/10.110…
“We also compare the performance of SARS-CoV-2 detection assays using seven different primer/probe sets and one assay kit.”
University of Washington (UW) RdRp primer/probe set
Corman N-gene, RdRp, and E-gene primer/probe sets.
Centers for Disease Control (CDC) N1, 91 N2, and N3 sets
Assay
The BGI RT-PCR detection kit (BGI)
I.e. the positive clinical sample was diluted by a bit more than 1000 times to give an undetectable sample.
Ten times more sensitive than the UW RdRp and Corman N-gene primer/probe sets.
“Assays using the CDC N2 and Corman E-gene primer/probe sets were more sensitive than those using the CDC N1 and Corman RDRP sets and the BGI kit”
This explains a comment UW virology made that they could reduce the amount of reagent need to extract RNA.